Hi,
I'm trying to count edges of an AI signal that are acquired with an NI-9221 on a cDAQ. The average frequency is about 1kHz.
I'm acquiring the signal with 10kHz but don't get the right amount of edges. 
Is
there any solution for this problem. I know, of course it would be
easier to use a CTR-modul but my signals doesn't fit with the specs.
Thanks for any help
Yves

Yes, I know. I wanted it to be posted in the Labview forum and hadn't realized my mistake.
But I still don't have any solution....
Yves

Similar Messages

  • USB 6008 Count Edges

    I am trying to count the edges on a signal with a period of 50us using the counter on a usb 6008.  Is this possible or is the period to small?

    Hi there!
    From the specifications listed at http://sine.ni.com/nips/cds/view/p/lang/en/nid/201986 and http://www.ni.com/pdf/manuals/371303l.pdf the minimum input pulse width is 100ns. Seeing that the period you are trying to measure is 50us, considerably bigger, then it should be fine.
    I hope this clarifies,
    Liam A.
    National Instruments
    Applications Engineer

  • Is it possible to use "Delay Values" to create a frequency signal from digital "Count Edges" -task? (= are the results I'm getting correct?)

    Hello.
    I have a digital encoder from which I need freaquency information (to ultimately get rpm -infromation). The problem is that this task is inside a loop with 2 other DAQmx -tasks that are using "one sample on demand"-aquisition mode and if I configure this new counter task to be a freaquency task, it only updates once in ~ second which makes the whole loop lag.
    I thus created an "artificial" freaquency signal by using "Count Edges" -aquicition mode and the "Delay Values" -block so that I substract the delayed signal from original "Count Edges" -signal. There is a 0,01s delay in the loop and I figured out that if the history of the "Delay Values" -block is 100 samples I would thus get the real edge-freaquency.
    I tested this configuration and the results seem to be at least really close to correct but I have no idea if this idea is in any way correct...
    This explanation was probably quite confusing so please see the picture attached.
    Thanks a lot in advance! 
    Attachments:
    are_the_results_correct.jpg ‏200 KB

    First of all, thank you for quick reply. Unfortunately I don't have the acces to the vi. until tomorrow.
    And yes, I think you understood correctly: essentially this arragement measures how many edges have been counted during one iteration. This is how I figured out that this could then be used as a frequency measurement:
    1. From the "Edges - Delayed Edges" I get the information on how many edges have been counted during last iteration.
    2. I "know" (really I don't?) that one iteration lasts ~0,01s because of the delay in the loop.
    3. There is 2048 edges in one round of the encoder so I get the rpm as follows: rpm = (edges - delayed edges)/204,8*60(s)
    (If I was using history size of one as you suggested it would be: rpm = (edges-delayed edges)/2048 * 60) However using history size of 10 and taking it account in the multiplication smoothens the response nicely.
    But doesn't this arragement count on the fact that the vi runs smoothly and there is no additional lag?
    I quess using another loop and notifiers for a dedicated freaquency measurement as you suggested could be worth trying. I just have to first learn how to use them. 
    If I do use them will the main loop run smoothly and not wait for every update of the notifier? This would be essential since the freaquency output refresses only about once in a second if I use the continious aquisition mode.
    Attachments:
    are_the_results_correct.jpg ‏198 KB

  • Timing problem with counting edges

    Hello!
    I use a NI PCI-6221 DAQ card with NI-DAQmx to count edges of TTL pulses for a spectrometric application.
    It is extremely important that I count the pulses for a well-defined period of time. Typically, I want to count the edges that reach the counter in the period of 400 milliseconds.
    Normally, this works quite well, when I use a WHILE loop that reads and restarts the counter every 400 milliseconds. Things change when the PC I run the VI on has other programs running in the background. Expecially computing-time intensive programs delay the 400 milliseconds of the WHILE loop for up to several 100 percent, resulting in an wrong counter read.
    I tried to use a timed WHILE loop but this din't change anything, regardless of the timimg source (onboard clock or PCI-6221 counter) I applied.
    Has anyone encountered similar problems and found a solution? Isn't there a possibility to control the counting time by hardware?
    Thanks in advance!
    EresthorMessage Edited by Eresthor on 04-07-2005 08:17 AM

    Hej Lynn,
    of course your idea works, thank you. It works extremely well this way, too.
    My problem with this approach is that I need both counters on the PCI-6221 to count data from my spectrometer simultaneously on two channels. That's why I would like to trigger the counter in a different way, for example with a clock.
    Is there no way to trigger the counter with the timebase of a hardware clock on the card? Or ist there another possibility without wasting a counter?
    Eresthor

  • Counting edges with dynapar encoder

    Hello,
    I am very new to Labview. To kick off some experience, I am trying to count edges on a Dynapar encoder, Model E14020000303. Here is the data sheet:
     https://ecatalog.dynapar.com/downloads/E14_DS_702489_2_.pdf
    I am using LabVIEW 2011, with a NI USB-6210 DAQ. My white wire on the encoder is connected to PFI0. The red and black wires are connected to my 5V voltage source. I am rotating my encoder to count the revolutions. Supposedly my encoder is a 200 PPR, and it is also quadrature. 
    However, I am not understanding what I am getting when counting edges using DAQ Assistant. I am using the Edge Count on ctr0. I can set the DAQ assistant to run on 1 sample acquisition mode and count the rising edge. When I turn the encoder 100 times, the measured value on the DAQ Assistant is usually around 76500, which would mean 765 edges counted per revolution. I can also change things around and go to continuous acquisition mode. I set up PFI2 as the external clock source with a rising edge and put the green wire in PFI2. I set it to read 1 sample at 1k Hz. This usually results in counting around 820 edges.
    I don't understand why I am getting these readings. The DAQ Assistant should only be counting rising edges because that is what it is set at. Why am I getting numbers so high? 
    Thanks!
     

    Hey guys,
    Sorry for the late reply, I have been put on other tasks. You all were right about the ground, that indeed was the problem. With the daq grounded, the encoder measures 200 edges per revolution.
    Thanks!

  • Count edges with hardware reset

    Hello all,
    I have a PCI-6220 and I would like to configure a counter to count edges with a hardware reset or use my counter to measure linear position without the B input, after words, only with the A input to count edges and the Z input to make the initial position.
    Somebody knows how?
    Thanks in advance, best regards,
    Paulo Carmo

    Duplicate Post
    John Passiak

  • Count edges on rtsi0

    Hi all,
    I have a PCI 7831-R and a PCI 6143 card running on my computer. Both
    are synchronized with a RTSi cable. The PCI7831 delivers via RTSI 0 a
    clock signal for Trigger my analog signals. My problem is now that i
    want count the edges on rtsi0 because I want to know what  the
    values  of my analog signal are refered to my triggersignal. For
    example, Trigger edge 1 - analog value x,  trigger 
    edge  2 - analog value x. Is that possible. I don't found nothing
    on ni examples. Thank you for your help.
    Grüße aus Deutschland
    Manuel

    Manuel,
    you want to count edges from a RTSI line with the PCI 6143, right?
    Have a look on the following post:
    http://forums.ni.com/ni/board/message?board.id=250&message.id=4715&requireLogin=False
    Regards,
    ThSa
    Message Edited by ThSa on 04-05-2006 09:42 AM
    http://www.newgistics.com

  • Alazar count edges synchronization

    Hello,
    We have an Alazar AT9350 of which the acquisition has to be synchronised with the acquisition of a Count-Edges channel on an NI-6259 DAQ board. Running them synchronously is easy, for we run the Alazar board with a trigger signal (50KHz) that we can route to the NI board as well. However, starting them synchronously poses more of a challenge. The original solution was to input the trigger on the NI-board and reroute it when the count edges VI starts. However, the board needs one counter to create the sample clock (or receive the trigger signal), and one counter to output the signal. This leaves no counter for the Count-Edges channel (our NI board has only two counters). 
    I have come up with a possible solution for synchronisation, but it is a bit ugly. I output the task just after the start task VI to an indicator, and create a local variable. I connect a read version of the local variable to the acquisition while loop of my alazar board. My assumption: because the while loop has to wait till all data flows have presented data to him, the while loop must start synchronously with the start task. The actual program is much bigger, that's why I'm using a local variable here. A direct wire is also possible, but it would look even uglier. 
    I have put a snapshot of a simplified version of my program as an attachment. I realise I am not actually saving any data or outputting what the Count-Edges reads, of course this is different in my real version. 
    Is my method viable or should I use a different method? (or perhaps it is viable, but I should still use a different method).  
    Solved!
    Go to Solution.
    Attachments:
    labview program.JPG ‏114 KB

    Dear GerdW,
    Thanks for your suggestions.
    [quote]Probably not "real time"… When you use DAQmx to read your input data you can synchronize input channels. Have you looked in the example VIs for that case? [\quote] I use an NI board with DAQmx drivers to acquire the Count Edges. I use an Alazar board with their own SubVI's for the other acquisition. Therefore I cannot use the DAQmx synchronisation (which would've been quite easy indeed). And indeed, by "real-time", I of course mean that the amount of data I process per second is the same as the amount of data that I acquire per second .
    I changed my example, hope I finally understood you well. I added a wait function to the loop that sends the notification. Otherwise the notification is sent on the same time, but there is still no synchronisation because the notifier is retrieved upon reaching the loop (and it essentially does nothing). To make sure I know when I start the two loops synchronously, I will add a high-resolution relative seconds to both loops and subtract them in another loop. If there is a better way to test synchronisation I'd happily take any suggestions. (Though perfect synchronisation is the best, one or two ticks off won't do much harm). 
    Attachments:
    MinimalExampleWhileLoopSynchronisation.PNG ‏14 KB

  • NI 5124 count edges

    Hi,
    Im using distance sensor ELGO EMIX3 (10-30 VDC square wave output). Max resolution is 0.01 mm while using 4 edge triggering.
    Current principle is to count rising edges from the signal (attachment)
    Problem is that I'm not able to count pulses as fast as needed. maximum speed for the sensor is 4 m/s and I'm nowhere near! I Would be happy for 0.1 m/s speed
    Is there a better way to count signal edges using NI 5124
    regards,
    Asmo
    Attachments:
    countedges.jpg ‏55 KB

    Thanks HSD,
    Code you provided does the trick and the system is able to count edges very fast.
    First I had to upgrade SCOPE driver software, because I got several errors about memory overload and one like this:
    > Component Name: nimxslu proxy
    > File Name: This is NOT an error in nimxslu. See nimxsl/tStatus2KernelProxyWrapper.cpp for information
    > Line Number: 290
    > Status Code: -218802
    Further, i tried to modify your code to count rising edges from two sensor signals with no success.(attachment)
    I added input channels 0,1 and get waveform array size 2. I also ensured that waveform array consist different signals from channels 0 and 1. Anyway, total records still show pulses from only one channel. I guess this has something to do with fetch records property node?
    and further..
    I need to count both rising and falling edges from two channel input. Sorry, Im not familiar using scope property nodes. Is it possible to modify this code you suggested earlier to count rising and falling edges?
    Attachments:
    countedges2.jpg ‏68 KB

  • Counting edges @ 90MHz

    Hi
    I'm looking for a solution to count edges from an external source without prescaler* @ 50 (90) MHz and read the counter (>=32bit) with a sampleclock of 100kHz.
    The M/X series Daq have counters  but the external sources routings limit the counter source to about 25MHz
    * I want every edge!
    Greetings from Germany
    Henrik
    LV since v3.1
    “ground” is a convenient fantasy
    '˙˙˙˙uıɐƃɐ lɐıp puɐ °06 ǝuoɥd ɹnoʎ uɹnʇ ǝsɐǝld 'ʎɹɐuıƃɐɯı sı pǝlɐıp ǝʌɐɥ noʎ ɹǝqɯnu ǝɥʇ'
    Solved!
    Go to Solution.

    John P wrote:
    Hi Henrik,
    I think you're talking about the TB-2715, which is a screw terminal block that mounts directly to the PXI-6602.  I'd want to actually set this up before giving a recommendation--the TB-2715 might work out but I haven't tried it before at high frequencies.  I'm hesitant on recommending a screw terminal connector block because I have had issues with the SCB-68 passing through high frequency signals cleanly, but perhaps the 2715 doesn't have the same problems since it connects directly to the 6602.
    Close I was thinking about that housing with the connector and a matching PCB core in Ultiboard for custom designs
    In the past I have used the BNC-2121 cabled directly into the SMB clock out connector of a PXIe-6545.  Both the output impedance of the 6545 and the characteristic impedance of the SMB to BNC cable were 50 Ohms.  I actually did not terminate at the BNC-2121 as I found it wasn't necesary with the cable lengths that I was using (I don't recall exactly what they were).
    The spec claims a 75Ohm input impedance for the 6602, well the mismatch probably don't hurt that much if you have short cables.
    The spec for the max external source for the 6602 only goes up to 80 MHz, so anything above that would not be a guarantee.  From past experience I believe it should work under the right circumstances, but it's probably something that would have to be set up to know for sure.  Do you have the contact info for your Field Sales Engineer who might be able to arrange for a loaner 6602 for you to try out?  Or, perhaps you can simply order one and return it if it does not suit your needs (check with your sales representative to confirm NI's return policy).
    I'm in close contact with NI here in Germany and a PXI-6602 was already offered for a test * 
    Best Regards,
    A maybe root idea, but how about using a PXIbus breakout board and feed the 90MHz to one of the internal triggerlines? Would that drill a hole through the f-max limited input?
    Spoiler (Highlight to read)
    The PTB seems to be a good value customer
    The PTB seems to be a good value customer
    Greetings from Germany
    Henrik
    LV since v3.1
    “ground” is a convenient fantasy
    '˙˙˙˙uıɐƃɐ lɐıp puɐ °06 ǝuoɥd ɹnoʎ uɹnʇ ǝsɐǝld 'ʎɹɐuıƃɐɯı sı pǝlɐıp ǝʌɐɥ noʎ ɹǝqɯnu ǝɥʇ'

  • NI PCIe-6351 Count Edges Channel error on fast TTL - Multiple Sample Clock pulses were detected

    Hello,
    I am trying to use a PCIe-6351 to record the arrival times of a fast TTL pulse stream (generated by an Excilitas/Elmer Perkin APD). The TTL pulses are 2.5 volt amplitude, 20 ns duration, with a gauranteed dead time of 50 ns between pulses. I am trying to use the the Count Edges function, with the  100MhzTimebase as the input terminal and the input to counter 0 (PFI8) as the sample clock. After a few seconds of acquiring data at 100 Mhz, the application throws the following error (-201314):
    "Multiple Sample Clock pulses were detected within one period of the input signal"
    I had thought that because there is 50 ns dead time between pulses, multiple pulses would never arrive within a single clock cycle of the 100 Mhz timebase. Is there any way this might not be the case? Alternatively, is it possible that the counter is triggering on some jitter around the edges of the pulses? If so, is there any way to filter such high frequencies without losing the 20 ns pulses?
    I have read through the forums for similar problems with photon detectors, but have not been able to resolve this issue. Thank you for the help.
    Matthew Bakalar

    It sounds like the input signal is being detected as multiple edges.
    The PFI filtering feature on the X Series card likely isn't going to be suitable for you.  The minimum setting is actually exactly 20 ns, which should in theory guarantee a 20 ns pulse passing through.  However, if the signal is high for anything less than that there wouldn't be a guarantee (depending on the phase of the timebase relative to the rising edge of the signal)--considering rise times and that there is evidently a glitch in the signal itself, it probably isn't actually a continuous 20 ns high time by the time the DAQ card sees it.
    What you should do instead:
    Configure a second counter as a retriggerable counter output (single pulse).
    Use your external signal as the start trigger for this counter output task.
    Set the initial delay, high time, and low time for the counter output task all to 20 ns (the minimum).
    Use the internal output of the counter output task as the sample clock source for the original edge count task.
    The counter output will be triggered when it sees the external signal, wait 10-20 ns, then generate a 20 ns pulse.  If there is a glitch on the trigger line during this 30-40 ns that the output is generating, it will be ignored.  The counter output will be re-armed in time for the next pulse given the minimum dead time of 50 ns between pulses.
    Best Regards,
    John Passiak

  • Losing Wi-Fi signal with cell signal simultaneously

    I'm in a crap area. I have 1 bar of 1x data where I am located. To counter that I have Wi-Fi. I'm the only user of the Wi-Fi and it's secured as well. My problem is my Wi-Fi connection on my phone disappears completely if I lose that one bar of 1x signal. I am able to reload my web pages while I have no signal with the tiny little no smoking sign at the top of my screen, but as soon as that one bar comes back, I lose all signal on Wi-Fi all over again!! It happens 1,324,501 times a day (that's a sarcastic count) but it leaky does happen extremely frequently. It happened 3 times just trying to your this on my phone. I've already deleted the Wi-Fi settings on my phone multiple times and reboot my router multiple times as well. How can I fix? ?

    My zip code is 28533. It works way better away from this area, I'm on a
    military base with no signal on it so I use Wi-Fi, which doesn't work with
    my phone. I've tried other Wi-Fi networks and they work fine unless I get
    low cell phone reception then the same problem starts to occur. My roommate
    has the same phone but with a different service provider and his phone
    works perfect with the Wi-Fi, as does my laptop and my 3Ds too. It's just
    my phone that acts foolish.
    On Nov 21, 2013 8:20 PM, "Verizon Wireless Customer Support" <

  • Better estimation of phase difference between two signals with variable frequency!

    Hello LabView Gurus, 
    Being a power engineer and having just a little knowledge of signal processing and labview, I have been pulling my hair out for the last couple of days to get a better estimation of phase difference between two signals.
    We have two analog voltage signals; 1. sine wave (50Hz ± 1Hz) and 2. a square wave with exactly half of sine wave frequency at any time.
    At the starting point of operation (and simulation/acquisition) both signals will have no phase difference. However, the square wave's frequency changes unpredictably for a just a few millisecond but then it gets synchronized with sine wave's frequency again. This means that the square wave will be phased out from its original position. The task of the labview is to find the phase difference between the two signals continuously.
    My approach to determine the phase difference is to measure the time when sine wave crosses zero amplitude and the time when the very next square wave changes amplitude from zero volts to +ve voltage (I have a 0.5volts threshold just to avoid any dramas from small line noise). The difference between these times is then divided by the time period and multiplied by 360 to get this phase difference in angles. 
    As this part is just a small block of a big project, I can only allow 5000Hz sampling rate each for both signals. I read 500 samples (which means I read data from 5 cycles of sine wave and 2.5 cycles of square wave).
    Now the problem is, as long as the frequency of sine wave stays constant at exactly 50Hz, I get a good estimation of the phase difference but when the frequency changes even a little (and it will happen in the real scenario i.e 50Hz ± 1Hz  and the square wave's frequency is dependent of sine wave's frequency), the estimation error increases.
    I have attached my labview program. From front panel, you can set the phase of square wave (between -180 and 0) and you should see the labview's calculated phase in the indicator box named 'Phase'. Then you can press 'Real Frequency' switch that would cause the frequency to change like it would in real operation.
    You can observe that the estimation error increases after you push the button. 
    All I need to do is to reduce this estimation error and make it as close to the actual phase difference as possible. Any help would be greatly appreciated.
    I am using LabView 2009 for this task.
    The application is for electric machines and the stability/performance of machines under different faults.
    Thank you for reading this far!
    Regards,
    Awais
    Attachments:
    v603.png ‏320 KB
    v603.vi ‏186 KB

    Jeff Bohrer wrote:
    Basic math gives me a bit of pause on this approach.  You are sampling at 50 times the frequency of interest so you get 50 samples per cycle.  your phase resolution is 1/50th cycle or 7.2 degrees +/- noise.  You will need to samlpe faster to reduce phase resolution or average multiple readings (at a time cost that is signifigant)
    Jeff- (Hardly Working)
    I am sampling at 100 times the sine wave's frequency and 200 times the square wave's frequency.  Increasing the sampling rate completely solves my problem. But since I am acquiring several other inputs, I cannot afford a sampling rate higher than 5kHz.
    F. Schubert wrote:
    I'm not a signal processing expert, but here my basic understanding.
    If you simulate sampling with 5kHz and a frequency of 50 Hz (and both are 'sync' by design), you always get an exact 5 periods. Any variation of your signals frequency gives you a propability to get 4 or 6 'trigger' events. That's an up or down of 20%!
    The one measure to reduce such problems is using 'window functions'. They don't fit your current approach (counting instead of a DSP algorithm), so this needs to be reworked as well.
    My approach would be to use the concept of a Locki-In amplifier. You need to phaseshift your ref-signal by 90°. Then multiply your measurement signal with the ref signal and the phase shifted ref signal. The obtained values for x/y coordinates of a complex number. Calculate the theta of the complex number (with the LV prim). Feed this in a low pass filter.
    The trick on this is, that the square wave has harmonics in it, in this you are interested in the second harmonic which is the sine wave.
    To get rid of the effect that the sync between sampling rate and ref signal frequency gives an error, you then can use the window I mentioned above (place it before the lock-in).
    For a design that really plays well, use a producer-consumer design pattern to get the calculations done in parallel with the DAQ.
    I suggest you to check on wikipedia for some of the keywords I mentioned. Go also for the external links which lead to great tutorials and AppNotes on the signal processing basics.
    Sorry, it's not a simple solution I offer and we will have quite some conversation on this forum if you follow this path. Maybe someone else knows a simpler way.
    Felix
    www.aescusoft.de
    My latest community nugget on producer/consumer design
    My current blog: A journey through uml
    An interesting view. the sine wave can indeed be looked as a second harmonic of the square wave. I will implement your idea and get back to you as soon as I get some results. But since I have very limited knowledge of signal processing, it might take me a while to get my hear around the solution you mentioned.

  • Dividing source signal with 2 counters

    I have a C/T 6602.
    I want to divide by n(>=2) one external signal with my two counters at the same time. But the problem is that I cannot make the two Counter outputs  in-phase whenever I set the counters up. Whenever the order is made, the two output phases are random. I think that there is way to do what I want.
    Thanks, GraceSun

    Hello GraceSun,
    Have you tried any example programs?  You have not said what programming language you are using, but if you are using LabVIEW 7.x you can go to help >> find examples >> hardware input and output >> DAQmx >> Generating Digital Pulses and find a VI called Gen Dig Pulse Train (either continuous or finite).  You can simply put in for your counter channel "Dev1/ctr0:1" to have multiple counters in the task.  Your outputs will then be synchronized. 
    Thanks,
    Laura

  • How to count area of the signal from the baseline?

    Hello!
    Please drive me to the function that can count the area above and below from the baseline of my signal.
    Like, i have the contineous signal, that has baseline,  and periodic oscillations. I need to count area of the signal tha different from baseline during this oscillations.
    Thank you.

    Assuming you want both the above and below areas to be counted as positive, take the absolute value of your function before integration.