Hi guys
I am using  a PCI 6289 M series device for A/D characterization , I am using the device both for the analog output generation and also for data acquistion of the digital lines , The dynamic tests for the A/D were showing a very poor SNR ratio of the ADC, it is a  12 bit adc so technically in the ideal case i should atleast see  90 db but the resulting figure was way lower around 40 db .
I connected the A0 from the PCI 6289 brought out to a scb-68 block to a RHODE and SCHWARZ spectrum analyzer , the above spectrum analyzer is a decent one with good analysis capability and I was surprised to see a SNR of 50 db , the data sheet says that PCI 628X devices have  a internal 16 bit DAC so technically the SNR should be atleast 90 db even if 6 db were removed for the harsh environment of the computer .
please advise , have anybody else faced this problem , the data sheet does not give a figure for the SNR ratio of the card , since i need to test a 12 bit adc , thought that 6289 would be up to the task since it has a 16 bit DAC.I am using the Noise shielding cables and also SCB-68 block , the block is far removed from other power supplies .
any ideas why the noise is occuring , one of my colleagues says maybe the harsh environment inside the pc enclosure causes it , is it necessary to use a separate PCI box for the card and then connect the card to the computer , is this standard procedure ?????.
and also does the PCI -6289 have a filtering option for the analog output , all i could find was the filtering for input.

Hi Tom
Thanks a lot for responding to the thread , did not expect a response in the holiday season . The SNR for the 12 bit one was a typo , was referring to the 16 bit onboard dac in the M series card , which technically should ensure a clean input signal having atleast 90 db .
The info you have given helps a lot , i suspect the conditions inside the pc might be causing it , also i am driving the analog output from the card at almost its limits of 2.5 MS/s , the sample rate specification for the card on the data sheet says 2.8 MS/s , also i notice the settling time requirement for the outputs is 3 us which i suspect will put the onboard DAC under quite a load.
If the conditions inside the PC cannot be rectified , where can i find a external PCI box which can be connected to the computer or does such a box exist only for the PXI cards ? will that necessarily rectify the problem and also since filtering is not possible on the card itself , I remember reading the scb -68 connector block has  single RC output filter option , i would have to use discrete R,C components to make it but notice on the right hand corner some pcb holes are provided , would this help.
Would appreciate your Response
Thank you
Chip learner

Similar Messages

  • How much noise expected in analog output?

    I'm using a PCI-MIO-16E-4 with LabView to generate buffered analog output, and at the BNC-2090 output (DAC0/1OUT) I'm finding ~180 Hz hum of about 20 mV peak-to-peak. It doesn't seem to matter whether I'm writing a waveform or not; the noise magnitude is always about the same.
    Is there something I can do to eliminate or reduce this, or is it typical so I'm stuck with it? I futzed with the BMC-2090 shield ground jumper (W1) but none of the three documented settings made any difference.
    Any suggestions would be greatly appreciated.
    thanks,
    Kevin

    No particular noise is "typical." No noise at all is atypical. 20mV is low enough that it could be radiated noise for nearby equipment. I doubt you card would have that noise in the lab situation where they test for noise. Which essentially means something about you estup is responsible for the noise.
    Since the frequency is a multiple of 60Hz it is probably environmental noise, radiated ot picked up off of the ground. Really and truly, 20mV is not a lot. I have acquisitions where I must be very careful in grounding ALL the associated equipment or I see huge blurts of noise from a motor drive - on the order of 3 Volts until I drove an eight foot gound rod through the floor to get an isolated safety ground for my DAQ PC and scope. I still see the noise, but it is now lower than 0.8V so I don't get false triggering of TTL circuitry from it. This is AI not AO like you are discussing.
    If you check the specification of your card there is a table on "System noise." I am not saavy enough to understand it, but I see that the noise is dependant on the output range. I suspect that the range is not a user selectable item, but in theory if you see more noise than the spec sheet says, then it is part of your setup not the card.
    Here is part you really don't want to hear: Solving a noise problem by text messaging is like solving a mental illness over the phone, maybe worse. I can tell you all sorts of practical, common sense things that are not specifically helpful. I can tell you from personal experience that you really need to follow all the good practices you can, if you really want to have as high a SN ratio as possible.
    Can you set up a single point ground? Can you eliminate all ground loops (shielding only grounded on one end.) I had a stray strand of shielding that brushed against a case that made terrible noise once.
    There is a good book called ,"Grounding and Shielding Techniques," by Ralph Morrison from Wiley Inter-Science. There are papers on this in in the tech resources of ni.com. In particular the paper on "Field wiring" is good. Keithley has a good book they will send you for free on low level measurements that is very informative. You have to call them on the phone to get it.

  • How do I configure 128 channels of DIO and 128 channels of Analog output with the availabili​ty of 3 pci slots

    I have only 3 slots of PCI or 4 slots of ISA bus, I want to configure with the minimum possible modules to attain 128 channels of DIO and 128 channels of analog output. This is inturn fed to the PCM (Pulse code Modulator). The cards will be placed in industrial PC.

    Associate Engineer-Marketing,
    Using just those 3 PCI slots you're not going to be able to get the 128 analog output channels. You will probably want to go either the MXI-3 (PCI bus extender) to a PXI chassis or E Series DAQ card to a SCXI chassis for such a high channel count system. Which analog output cards and which digital I/O cards you will want depends on your application. You will have to decide whether you will need static or dynamic analog output as well as whether you need clocked digital I/O or not. I highly recommend contacting our internal sales department at 1-800-433-3488. They are great at helping specify systems.
    Ames
    Applications Engineering
    National Instruments

  • PCI Card for Analog output in the range of 10mv

    dear ni,
                    I want to know about the PCI card that can be generate analog output in the range of maximum 10mv. i  am going to use for caliberation of loadcell, strain gauges devices.
    could you tell me on which PCI card will support this type of application.
    Regards,
    Balaji DP

    Try:  http://www.ni.com/dataacquisition/
    These have analog output voltages < 10V:   http://sine.ni.com/nifn/cds/view/main/p/sn/n12:7604,n3:7853/lang/en/nid/1036/ap/daq
    You need something with a high bit count to get good resolution at 10 mV, such as the PCI-6010 which has a 16 bit D/A.   
    Here are the minimum voltage specs for the 6010:
    Minimum Voltage Range
    -0.2..0.2 V
          Range Accuracy
    283 µV
          Range Sensitivity
    6.4 µV
    Message Edited by vt92 on 11-18-2009 07:56 AM
    "There is a God shaped vacuum in the heart of every man which cannot be filled by any created thing, but only by God, the Creator, made known through Jesus." - Blaise Pascal

  • I am using an NI PCI 6024 E board device for an analog output application. Can I generate two independent waveforms, one on each analog output channel?

    I am using an NI PCI 6024 E board device for an analog output application. Can I generate two independent waveforms, one on each analog output channel? In attach I send a example of my vi to control one chanel, but i need to control both but with diferent waveforms.
    Is the vi correctly programed to this propose?

    Yes you should be able to do that. look at the signal generation examples shipped with LabView
    You can create a 2 D array, with each row representing a wave form per channel. The number of points of the wave form (per row of the 2 D array) represent the number of points in the output buffer
    The wave forms can have different amplitudes/shapes, and therefore they are independent in this sense.
    However, you need to set the update rate,which is the same for both channels. The update rate together with number of points per buffer determines the frequency of the wave forms. This means the two wave forms will have the same frequency.
    To have different frequencies, you need to have say, on wave form with one cycle per buffer, and the other waveform have 2 cycles
    per buffer. in a case like that the frequency of the second channels is twice that of the first channel, and so on
    The two wave forms are then not truly independent, they may have different amplitudtes/shapes, but related in frequency.

  • Setting different analog output's parameters in PCI-6229 DAQ card

    Hello,
    We are using DAQ assistant to set the analog output parameters (frequency).
    Is it possible to set different values for each channel using DAQ assistant?
    If not using DAQ assistant, what  is the other way to set it?
    Thank You
    Arunas

    If you are trying to set different frequencies for waveform generation on two channels, the answer is No, it can't be done. If possible, you will have to pick a sample rate high enough to work for both signals, and use an over-sampled waveform.
    Another choice would be to use a second device for the second waveform.
    John Weeks
    WaveMetrics, Inc.
    Phone (503) 620-3001
    Fax (503) 620-6754
    www.wavemetrics.com

  • Is inactive analog output will be at high impedance or cut off state?

    Hey 
    I'm using the NI 6229 analog outputs.
    I would like to know if there is a tri-state or high impedance state for those  analog output when it is not in use? 
    Solved!
    Go to Solution.

    Hello amco,
    From the description of your project, I'm envisioning two probes attached to different points along a single trace on a PCB.
    It is certainly not recommended to have two analog output lines attached to the same channel.  As you have experienced, the behavior is unpredictable.  In a worst-case-scenario, one could potentially damage the AO line that is not actively driving the channel.
    So we do need to find a configuration that makes the unusued AO channel electrically separated from the channel-under-test.  One option would be to introduce an electromechanical switch.  This will add some complexity to your setup and your code, but it will permit you to send the signal from AO-0, disconnect AO-0, connect AO-1 and send a signal from AO-1 — all in a very short amount of time.
    Best regards
    Matthew H.
    Applications Engineer
    National Instruments

  • When using the analog inputs and analog outputs of the PCI-7344, what is the conversion between the voltage entering the card to counts? Similarly, what is the conversion between counts to voltage at the output of the card?

    I am using the PCI-7344 to control my system. The analog inputs are connected to the output of my system while the analog outputs serve as feedback to the system. The system is a servo. I want to know what is the conversion between the voltage read at the input, to card counts, and finally to the voltage output to the system.

    Carole,
    If you are trying to do analog feedback with a servo motor, Chapter 14 of the manual talks about how to set up the torque feedback. Also linked below is a LabVIEW example of analog feedback.
    Chapter 14
    NI-Motion User Manual
    Automatic Analog Feedback with FlexMotion example
    A. Talley
    National Instruments

  • How do I get an analog output signal from my PCI-MIO-16E-4 card?

    I need an analog output from 0-10 V to control the mass range on a mass spectrometer. Do I use the DAC0out channel to do this? The manual only talks about analog out (ao) channels. Is this the same as DAC0out and DAC1out? How do I set up the DAQ assistant for this process? I do not need a waveform, only a constant signal.

    Yes indeeded. The DAC0out and DAC1out are the analog output channels (ao0 and ao1).
    Simply follow the examples that ship with LabVIEW. You can also easily create code with DAQ assistant.

  • Synchronize two PCI Board for Analoge Output

    Hi All
    Has anyone a good example for synchronizing two MIO Boards over RTSI cable.
    I found a lot examples to analoge input but nothing about analoge output.
    Thanks

    Hi Reto.
    I posted to this once, but looks like it didn't make it. I checked the examples, and Ir eally didn't see anything that did what you wanted over the RTSI bus, so I made one. Check it out.
    Mark
    Attachments:
    RTSI_AO.vi ‏95 KB

  • Combine Analog outputs for higher current drive

    I'm using a cRIO 9263 analog output module.  The current output specification is 1 mA.  If I need 1.2 mA of current, can I use two outputs in parallel?
    SteveA
    CLD
    FPGA/RT/PDA/TP/DSC

    I am not familiar with that device, but, in general when paralleling voltage output devices which were not specifically designed to be paralleled you need to include some ballast resistors. They allow for slight differences in the output voltages of the two (or more) devices and promote current sharing.
    Put one resistor in series with each output. Connect the other ends of the resistors together and connect this to the load. The value of the resistors depends on the amount of current, the voltage difference between the two outputs when given the same digital code, how much voltage drop in the resistors can be allowed, and perhaps other factors such as how fast the output must change.
    I can help with the calculations if you are not familiar with them, but I will need more information about the voltages and loads.
    Lynn

  • How to building a multi-channels analog output task in visual c++ 6.0 (without Measurement Studio)?

    Hello!
     I have a PCI 6251 card, and using DAQmx C function to generate a wavwform. (single channel ).
    But, how to creating a multi-channels analog output task, and had a different frequence in each channels?
    Thanks.

    This is getting to be a common question!
    The answer is, you can't.
    If the rates have common factors, you can use a higher overall rate. The slower waveform will simply have extra points in it to make up for the higher update rate.
    John Weeks
    WaveMetrics, Inc.
    Phone (503) 620-3001
    Fax (503) 620-6754
    www.wavemetrics.com

  • Pci-6289 delay

    Hi guys
    I am facing a problem with simultaneous clocking and triggering capabilities of the PCI-6289 Multifunction Daq card. To facilitate easier understanding the application i am working on is bench characterization of adc using the PCI-6289 , I use the 16 bit D/A onboard the card to generate the input signal to the adc , each time a conversion happens the adc puts out a pulse signal , I use the pulse signal as an clock for the data acquistion to happen exactly each time an conversion is done.
    The thing is I want the signal generation and data acquistion to be synchronized perfectly , have tried a couple of variations of programs in Labview .I have attached two versions of the programs , one uses the more basic data acquistion functions to generate the signal and try to acquire it at the same time after averaging out some samples , the other version tries to use the daq assistant to try and synchronize the data acquistion with the generation .
    In each version there appears a small delay which translates to code offset in the very begginning , The daq assistant grays out the triggering area which i suppose could be used for making sure the data acquistion only happens when the signal is present and only when the conversion pulse is high.
    If anybody could suggest a way for the clocking and triggering using either daq assitant or the more basic vi's ,it would be great.
    Also if it could be clarified if this is a limitation of the card would appreciate it.
    thank you
    Chiplearner
    Attachments:
    datavalid_validl12.llb ‏1848 KB
    lv gen.vi ‏474 KB

    What you need to do is perform Multi function Synchronization
    Read this link to understand Multi function synchronization
    Also search in Examples shipped with LabVIEW to find Multi function synchronization example VI's
    Find examples>>Hardware input output>>DAQmx>>Synchronization>>Multi function folder
    Message Edited by devchander on 08-23-2006 03:27 AM

  • How to avoid jitter in analog output(square wave) generation

    Hi all,
    Iam working with PCI-6723 32 channels analog output,iam generating a square wave,sampling at 100ks/s.
    ican able to generate upto 10kHz square wave.The wave form appears clean when it is below 1000Hz if i crossed 5KHz to 7KHz, iam getting jitter in the output square wave form,i have experimented with the digital filter but still the jitter is persisting,
    since iam using 12 channels for such similar generations so i have fixed the sampling rate.100Ks/s.
    kindly suggest me how to avoid this jitter in high frequency generation.
    OS::WINDOWS 2000
    HARDWARE PCI-6723.
    I/O CB 68
    NIDAQ:7.3
    LABVIEW 7.1
    kindly find the attached file.
    Thanks and Regards
    RAJESH
    Attachments:
    Untitled.vi ‏73 KB

    Hi ALAN,
    Thanks for your reply, well the square appears to be overlapped ie one ON/OFF period another waveform tries to overlap, actually i planned to take a snapshot of my oscilloscope (tetronix) but since its not advanced one ,i have attached the file kindly look into it,you can observe the at some higher frequencies there is lot of difference between rise time and fall time.
    Thanks and Regards
    rajesh
    alan actually iam decorating my front pannel do u have any sample layouts for coloring ,buttons kindy forward me .
    Attachments:
    Doc1.doc ‏56 KB

  • How to avoid delay during analog output generation by changing its frequency?

    Windows XP
    LabVIEW 7.1
    PCI-6036E + BNC-2120
    Hi,
    I am going to create a vi to generate an engine speed sensor signal (a simple square wave with specific missed pulses, in my case 58 pulses “teeth” and 2 missed pulses “missed teeth”) as an analog output but in addition give me the opportunity to control parameters for example frequency online to simulate the engine speed changes during running that vi. For this purpose I have started with “Continuous Generation.vi” which is available in NI Example Finder under the following path:
    Hardware Input and Output > Traditional DAQ > Analog Output > Continuous Generation.vi
    Then I modified it towards above mentioned goal, all related vi s are attached. The main vi is: "Motor Signal Generator_1.12.vi"
    At the first try it looks that it works properly but when have a look on that more accurately with Oscilloscope (fortunately I have a good one: Agilent 54621A – 60 MHz, 200 Ms/s) obviously there is a gap (delay or Jitter) whenever I change the engine speed. It is also attached in Signal generation_problem report.doc file.
    Note: Small gaps are OK and related to predefined missed teeth but the big one is happened during changing engine speed.
    As far as I understand it is related to the time which case structure in AO C-GEN sub-vi needs for AO reconfiguration each time after changing the engine speed (update rate). How can I get rid of this delay or gap during signal generation and generating completely continuous signal?
    I have to mention that obviously I changed the frequency by changing the update rate. The other possibility is to change the number of updates in one period (refer to "generate arb frequency.vi" in NI site: http://sine.ni.com/apps/we/niepd_web_display.display_epd4?p_guid=B45EACE3E48F56A4E034080020E74861) which resulted in no delay however then I can not change the frequency continuously but step by step (for example jump from 5Khz to 2.5KHz immediately) and this can not pass to my application.
    Any hint is appreciated.
    regards
    Attachments:
    Signal_generation_NIsupport.zip ‏81 KB

    Hi Roozbeh,
    The following example will allow you to vary the pulse train frequency during run time.
    Thanks,
    Lesley Y.
    Attachments: