Hi,
We have a NI-4461 Dynamic Signal Analysis card.I need to plot a Freq Spectrum & Bode Plot(Gain Vs Freq & phase Vs Freq).I am able to acquire data using DAQmx functions,but the DAQmx library does not support Signal  Analysis functions such as FFT,gain and phase plotting.I would like to know if a DSA library is available for this card and how do I download it and use it with the card?
Thanks in Advance
G.Bhaskar

Hi Bhaskar,
There's a difference between studying a signal's frequency response and a signals frequency spectrum. For example if you are interested in the response of Vo/Vi as the frequency changes, you would use a Bode plot. A Bode plot is the amplitude vs. frequency or the phase vs. frequency. Usually frequency is logarithmically scaled and the two plots are stacked so you can see their relationship to each other. Phase will change with frequency as well as the amplitude. This is useful in studying transfer functions and how circuits respond to different stimuli.
If you are interested in the spectrum of frequencies that are contained in a signal, you would use an FFT. The highest amplitude would correlate to the fundamental frequency of your signal. This is not the same as phase. If you take an FFT, you will be specifying the frequency resolution (which may not be apparent but your parameters and array length, amongst other things, will set your resolution). So if you're resolution is 10Hz and you're FFT(Vo) = [.2, .2, 51, .2, .4, .3], your fundamental frequency would be 30Hz. Any other peaks above the noise floor will represent harmonics or other tones in your signal. A musical chord, for example, could contain 3 tones but many harmonics.
Please take a look at these documents
The Fundamentals of FFT-Based Signal Analysis and Measurement in LabVIEW and LabWindows/CVI
http://zone.ni.com/devzone/cda/tut/p/id/4278
An Overview of Frequency-Domain Measurements
http://zone.ni.com/devzone/cda/tut/p/id/3011
PBear
NI RF

Similar Messages

  • The Signal analysis functions of the DIAdem-CAL can not use.

    I use a DT3003 board for Data Acquisition in my system. My data is signal of Electromyography,and it needs the DIAdem-CALC "Signal Analysis" such as FFT.But these functions in my DIAdem are disable, Why?
    Attachments:
    10min.rar ‏194 KB

    Yes!
    We do need Frequency Analysis.The EMG is some kind of time-voltage signal.DIAdem do us great help in data sampling.We need to get information such as Average Amplitude, spectral parameters like Mean Power Frequency,Median Frequency from the raw EMG. Is "Signal Analysis" function of the DIAdem standar version able to full fill our demands?
    Re:Thanks for your advice again.Your helps are important for us.
    Cediland

  • On the block diagram, clicked Express - Signal Analysis, I don't see Statistics​.

    On the block diagram, I clicked EXPRESS - SIGNAL ANALYSIS , then all I get are SIMULATE and HISTOGRAM. What happened to the rest like SPECTRAL, TONE, STATISTICS, etc.?
    Solved!
    Go to Solution.

    (Well, if you get a sine wave, you are probably simulating one. )
    Where do the temperature readings come from? Please show us your actual VI. This seems almost trivial.
    LabVIEW Champion . Do more with less code and in less time .

  • Signal analysis for DAQ

    Hello,
    what is the easiest way in performing signal analysis on a signal from a DAQ device?
    For example if i can obtain a sine wave signal from a DAQ ,how can i program labview to give me a indication if the signal is correct ie. sine wave, ampltude, frequency, etc etc
     with the option of setting the error of eg. 3%
    and also if the curve is unorthordox, can you plot points if the ideal signal, to determine and compare it to the signal obtained from the DAQ and say if it is ok ? with the option of error.

    Hi Labviewwuser3,
    Good afternoon and I hope your well. Thanks for your post.
    You have many options to analyses your signal.
    The easiest is maybe an express VI called Equal, it allows you test if two waveforms are the same, greater, less and within tolerance (error).
    More in depth you could use some of the VIs in the maths>> Signal processing >>Signal Operation such as peak detection. You could then compare the locations of peaks and amplitude. 
    You could also do this on a point by point basses or on a complete waveform...
    Hope this helps to get you started, 
    Message Edited by Hillman on 12-18-2008 01:31 PM
    Kind Regards
    James Hillman
    Applications Engineer 2008 to 2009 National Instruments UK & Ireland
    Loughborough University UK - 2006 to 2011
    Remember Kudos those who help!

  • How lab view can be used for signal analysis of hf signals recrded in field

    im interested in using labview tool to analyse hf signals in terms of removing noise ,amplipying signal strand identify modulation schemes used

    The kind of things you mention are certainly within LV's capabilities. When you say "HF" what frequencies do you mean?
    Mike...
    Certified Professional Instructor
    Certified LabVIEW Architect
    LabVIEW Champion
    "... after all, He's not a tame lion..."
    Be thinking ahead and mark your dance card for NI Week 2015 now: TS 6139 - Object Oriented First Steps

  • Signal analysing program faces different signal processing errors while the mentioned settings in the error solutions are correct!!

    I am trying to get 4 signals from my bluetooth  which is connected to a sensor. later on I need to analyze the signals for calculating heart beat and blood oxygen saturation. when the labview reads the data it has no errors but anyway the peak detecor doesnt work. my sampling rate is 100. anyway I get different signal processing errors like 20023 and 20037 and 20041, 20015. but when i check the solutions for the errors, they dont help me at all. like error 20023 complains about low and high cut of frequency while mine is set correctly could you please check it for me
    Attachments:
    shift final - Copy.vi ‏458 KB

    Mistaken wiring here- after BD Cleanup (Ctrl-U)
    Error -20037 occurred at an unidentified location
    Possible reason(s):
    Analysis:  The number of data points in the Y value array must be greater than the order value.
    Jeff

  • Cost effective video signal analysis

    Is it possible to get LabView to perform analysis on a video signal (NTSC) and if so, what devices (DAQ cards, video capture cards, etc.) do I need to look at. We are looking at options to improve our video quality testing at our factory and are looking for a cost effective soultion. There are video signal analyzers out there, but I would like a PC based soultion since the factory is beginning to use LabView for other testing. Any advice would be helpful.
    Troy

    Troy. National Instruments has an entire line of machine vision and image processing hardware and software. If you're processing an NTSC signal, I recommend an NI-1411 color frame grabber. From there, you can use the NI-Vision software for LabVIEW to analyze your NTSC signal. Along with NI-Vision for LabVIEW, I recommend using a product called Vision Builder, which is a prototyping environment that lets you play with most of the Vision fuctions before even openning LabVIEW. Then once you're done prototyping, Vision Builder will create your LabVIEW VI for you!
    Please respond if you have any further questions.
    Kyle V
    Applications Engineer
    National Instruments

  • EMG Signal Analysis

    http://sine.ni.com/apps/utf8/niae_asc.main
    I have a query regarding amplification , filtering of the EMG signal directly connected to PCI 6221 DAQ without any external signal processing tool. Also EMG ranges upto 15 mv and usable upto 300 Hz. More you can refer the above link

    Hi kapillohia,
    To address your question on acquiring an EMG signal, there are a few things we need to take into consideration.
    On the hardware side, I understand you have a PCI-6221. As John mentioned before, you will be able to feed the signal straight into an analog channel and see data. However, the resolution will be better if you did some signal conditioning and amplified the signal before connecting to the DAQ card. For more information on this process, you can look around this article.
    In LabVIEW, you will use the DAQmx application programming interface (API) to communicate with the hardware to get your data. John walked you through on how to use the DAQ Assistant which is the DAQmx express function. There are many more examples in the LabVIEW Example Finder which can be accessed from the Help menu. In there, you can find relevant examples browsing by task under Hardware Input and Output » DAQmx » Functions » Analog Measurements » Voltage.
    Feel free to look at a this other link to get an idea of how a similar application may be developed. Let me know if you have any specific questions.
    Regards,
    Kent
    Applications Engineer
    Digital Multimeter Home

  • Ecg SIGNAL ANALYSIS

    please kinly help me. i want to detect the heart rate from the ecg signal. i am using labview 8.2.1 student edition

    You may also want to have a look at the Life Sciences forum located here
    You can post additional questions within that forum.
    Here are additional threads to look at:
    LINK 1
    LINK 2
    R
    Message Edited by Ray.R on 03-16-2009 12:34 PM

  • Stepped signal analysis

    how to count the number of steps(red signal on graph) on the diagram below?
    number of steps should be counted in the rising and falling signal and make a comparison between them.
    how this can be done?
    Attachments:
    diag.jpg ‏55 KB

    There are probably lots of ways.  Your diagram has a hint for one of them.  Here are a couple of options:
    Use a Savitzky-Golay filter to take the derivative of the waveform.  You will end up with a bunch of peaks and valleys.  Find the peaks with a peak finder and count them.  You can figure out up/down by looking at the area between the peaks and get an average level or by looking at the direction of the peaks (up is up, down is down).  You data is fairly noisy, so you will have to experiment with how wide you want the S-G filter to be and how much averaging you want to do between the peaks.
    Use a single cycle square wave as a convolution filter and run it.  The end result will be similar to the Savitzky-Golay filter and can be analyzed the same way.  This is probably easier and faster on some targets, such as an FPGA.
    Good luck.  Let us know if you need more help.
    This account is no longer active. Contact ShadesOfGray for current posts and information.

  • Independent 7 Year LabView (3.0-7.1) and Real-Time Programmer w/ Signal Analysis and Audio/Video/Imaging Background

    I have been a LabView programmer since version 3.1 and now into version 7.1 - including the Real-Time platform. I also have developed database connectivity in Perl/CGI/HTML in conjunction with this experience.
    I also have over a decade of experience in Audio, Video and Imaging technology as a systems designer and production engineer. I am currently a part-time employee and an independent contractor. My company name is The Oakland Group. I am available for both short-term full-time and long-term part-time projects.
    I look forward to discussing any LabView project with you.
    Tom Held
    [email protected]
    414-964-0518

    Please see my attached resume.
    Tom Held
    Attachments:
    resume-scada.pdf ‏43 KB

  • Noise measurment of an accelerometer with the PCI-4461 card

    Hi everyone, 
    I'm trying to measure the noise of an single axis accelerometer with a PCI-4461 card. 
    First, I have manually measured this noise using an HP35665 signal analyser. I obtain something like this: 
    And now with LabView and the PCI-4461 card, I get this: 
    My question is: where does this part come from? And how to remove it ? (Since it doesn't exist with the signal analyser)
    I'm using LabView 8.5.
    This is how I configured the DAQ acquisition:
    and this is my VI:
    Thank you in advance for helping 
    Arthur
    Solved!
    Go to Solution.

    Hello 
    Yes indeed, the time signal seems to have two paterns (in green below), once negative and positive once about 50 ms later. This pattern is repeated with a frequency of 0.94 Hz (for simplicity let's say 1Hz).
    I have reproduced this pattern with matlab using a decreasing exponential:
    I have obtained this spectral signal:
    That closely looks like what disturbs the spectrum, indeed..
    Then I have tried to output a sinewave and acquire it:  
    I didn't see any problem.
    So I made the measurment with other sensors.
    The sensors are ICP accelerometers.The sensors that I am using are in page 7 of these document: 
    https://docs.google.com/viewer?url=http%3A%2F%2Fvibrasens.com%2Fimages%2Fpdf%2FVS_Catalog.pdf
    (I'm currently doing an intership in this company) 
    And finally I found that the problem came from the cable: when I use the shielded cables, the problem is present; and when I use "normal" cable, there is not the problem, as you can see :
    (In red, the sensor is connected with a non shielded cable, and in white, with a shielded one.)

  • Signal Generation through pt by pt Signal

    Hello,
    I am acquiring signal through serial port, to analyze the signal, i am using pt bt pt signal analysis.
    In order to work on it, i used sine wave pt pt signal generation. But remained un successful in generating sine wave through it.
    Kindly guide me how could i generate sine wave using pt by pt.
    Thanks

    Hi Intelligent,
    it's not a good idea to start asking by using PNs, I will ignore them in the future!
    You wrote:
    I already read the help and tried to solve the issue but after not receiving the satisfied results, i posted on the group. As its written in the help that the frequency is in Hz, but the value you have selected for the frequency is 0.01. So its un understandable, also the dependentence of time factor.
    - What are "unsatisfying results"?
    - Yes, the frequency is given in Hz and I chose 0.01Hz for sake of showing an example...
    - "Dependence of time factor" is explained in the context help. Did you really read the help? Or do you have problems to understand the simple math formula provided in the help? All the parameters of SineWavePtByPt are used in that formula...
    Best regards,
    GerdW
    CLAD, using 2009SP1 + LV2011SP1 + LV2014SP1 on WinXP+Win7+cRIO
    Kudos are welcome

  • Noise, Vibration & Order Analysis

    Dear all,
    I am currently looking for a program that can perform a noise, vibration and order analysis. I was just made aware that LabVIEW (Developer Suite, if I'm not mistaken) comes with Noise and Vibration and Order toolkits. I've read through the data sheets for both toolkits. Now, I'm wondering whether the two mentioned toolkits can perform the analysis below:
    1. Vold-Kalman Order Tracking Filter
    2. Modal and Structural Analysis
    3. Noise Source Identification
    4. Modal Test Consultant
    5. Acoustic Test Consultant
    6. Real time CPB Analysis
    7. Rigid Balancing and Two-Plane Balancing Consultant
    8. Baseband FFT Analysis
    9. Signal Enhancement
    10. Auxiliary Parameter Logging
    11. Cepstrum Analysis, and
    12. Envelope Analysis
    These analysis modules can be performed by Bruel & Kjaer product. I'm trying to influence my superior that National Instrument can also achieve the same objectives. So, maybe the users out there, or NI technical experts can help me with this. If the two toolkits are insufficient to perform such tasks, but NI has other toolkits that can supplement, can you guys please mention it in your reply.
    Many thanks!
    Shazlan

    The order analysis tools and the sound and vibration tools provide the fundamental signal analysis for most all machinery vibration functions.   However, the application level such as hammer blow test (modal), balancing, and so forth are not complete solutions. 
    I recommend looking at the following solution, which is based on these tools and comes with a turn-key application solution....
    www.vibdaq.com
    Let us know what further questions you have.
    Preston Johnson
    Principal Sales Engineer
    Condition Monitoring Systems
    Vibration Analyst III - www.vibinst.org, www.mobiusinstitute.com
    National Instruments
    [email protected]
    www.ni.com/mcm
    www.ni.com/soundandvibration
    www.ni.com/biganalogdata
    512-683-5444

  • Looking for a UK based Labview programmer, short term, contract or full time, depending on what suits. Immediate start, Signal Processing skills ideal.

    We are a UK South-West based LabVIEW consultancy company looking for a LabVIEW programmer preferably with advanced signal analysis and processing skills, to become part of the team or work as a short term contractor. We will be looking to advance their skills through a range of different projects that we work on, from aerospace test and measurement, through to shop floor data collection and User Interface marketing.
    Must have at least two years experience in LabVIEW and have an electronic engineering based degree or be able to show signal analysis is an area of expertise. We are looking to fill the position as soon as possible.
    Please your CV’s to the below address.
    [email protected]

    If the candidate had all the correct attributes then it could indeed be an option.
    Regards
    Peter

Maybe you are looking for