Hey guys.
I have a problem with my peak detect on a Power spectrum signal. from http://www.ni.com/white-paper/4278/en/ I know that 
I would like to have f = 1 meaning that the frequency at which the acquired time-domain signal was sampled (Fs) has to be equal to the number of points in the acquired time-domain signal (N). 
When i try to do this the peak detect returns the error 20013 the width must be greater than zero and smaller than sample rate.
this makes no sens to me ? 
When i have Fs = N x 10 it runs. 
I have added the VI that i have problems with ! if you need the whole project I will gladly add that to.
Attachments:
Both sensors.zip ‏329 KB

Hi Genex,
instead of that huge VI you should attach a smaller one, that demonstrates the problem! Right now we would need your DAQ setup to generate a signal to be able to try that VI...
So please:
Generate some typical waveforms, make them default in a waveform control and attach a VI with those waveforms and a typical call (with all your parameters) of PeakDetector!
Best regards,
GerdW
CLAD, using 2009SP1 + LV2011SP1 + LV2014SP1 on WinXP+Win7+cRIO
Kudos are welcome

Similar Messages

  • 6i can't find waveform peak detection vi. Load error code 20

    6i can't find waveform peak detection vi. Load error code 20

    The waveform peak detection VI is found only in the Full and Professional development systems. Do you have the base package?

  • How to measure error on peak detect?

    Hello!
    I'm using the vi Waveform Peak Detect to find peaks in some signals that we analyze at the lab where i work. However, i must know also what's the error on the measure. Anyone knows if there is a vi that measures peak AND the error on the values found? Or at least a mathematical way to find the error myself?
    Thanks!

    This is not an simple question to answer. When you search for the peak value of a signal you have to make some assumptions about that signal. For example if your signal looks like [1, 1, -1, -1, 1, 1, -1, -1, ...], it may be difficult to define its peak value.
    The first approach is, that your signal is a square wave of amplitude 1.0 and normalized frequency = 0.25. It this case the peak value is also 1.0
    But the same signal could be the result of the sampling of a pure sine tone of amplitude 1.4142 (SQRT(2)). if you sample your sine tone at 45, 135, 225 and 315 degrees of phase, you get the same sequence of samples. In this case you would like your peak result to be 1.4142.
    So the error can be as big a 41% (3 dB). This is of course an extreme situation, and
    in most case the error is much smaller.
    In general, the higher the relative bandwidth of your signal is (the ratio between the signal bandwidth and the used sample rate), the most uncertain is your result. So if you want to "stay in control" you need to make sure your signal is sampled at a rate that is high enough to give a good representation of your original analog signal.
    As an option, you may want to re-sample your signal (to a higher rate) using an appropriate algorithm. The new resampling VIs offer various interpolation algorithm that you can experiment with, depending on the nature of your signal. The peak detection VI will then do the work for you.

  • Fetch and peak detection all channels of PXI-5105 with 4M record... HELP!

    Dear collegaues!
    Please help me to improve performance my application, see attachment, and sorry for my English.
    So, my task is to fetch and peak detection all (eight) channels of PXI-5105 with 4M record and sample rate 4M with loop 1 sec...
    Inputs of all my channels are wiring to NaI detectors with 0,5...1 microsec of pulse width (really) and freq from 0 kHz to no more than 40 kHz.
    Why I've selected 4M record and 4M sample rate namely? Answer is that I've tested PXI-5105 previously by generator 40 kHz and 0,5 microsec width pulse. It is working fine and peak detection indicate for me 40000 pulses/sec. If I set lower than 4M record and 4M sample rate it is no working. In my honest opinion 4M record and 4M sample rate are very min settings.
    In the present time peak detection working only 6 channels... When I've connected to diagram more than 6 "peak detector.vi" - I see the error "...out of memory...".
    Please advise me, what is to be done for that is all working fine.
    Solved!
    Go to Solution.
    Attachments:
    consumer-producer7.vi ‏44 KB

    What you are running into is an out of memory error in LabVIEW.  You have enough onboard memory to capture 4M samples per channel on the digitizer.  The issue is with fetching and manipulating that data in your LabVIEW application.  You will want to step back and take a look at how you are handling your data to understand why that is happening.
    1) 4M samples/ch = 4M Samples x 2Bytes/sample/ch = 8M Bytes/ch
    2) Expanding to 8 channels creates 64M Bytes of data in the raw binary format
    3) You are scaling your data by fetching in a 1D WDT format.  This stores each sample in a 32 bit double, expanding the memory to 256M Bytes (in addition to timing information)
    4) By splitting up the array of waveforms and branching data it you can easily create copies of this data, and if your consumer loop is not completed with the last data, you may be trying to capture a whole new set, creating yet another copy.
    So you can see that while you have 1.5GB of controller memory, when dealing with large arrays of data you can easily eat up that memory.  There are several things you can try to make your application more efficient.  You could work with an unscaled binary data format, you can wire the array of waveform directly to the peak detect vi (instead of creating 8 copies, you will have a single copy with arrays of output) or you could revisit the record size you have chosen (experimenting with your threshold and width settings might help you to get the results you want with smaller record lengths).
    -Jennifer O.

  • Waveform peak detection ...Please have a look

    Hi folks,
               I am trying to acquire 2 signals and then detect the peaks of each of them and divide the peaks and display it in a waveform chart.But my Peak detection doesnt work at all...I tried plugging in values from 3 to 10000 but it just doesnt work...Please have a look at my VI and feel free to point out any errors in my VI..
    Thanks
    Attachments:
    slope 2 channels3.vi ‏51 KB

    Hi,
    Had a look at ur VI
    tell me, do u need a single peak for every iteration or a series of peaks above a certain threshold?
    If it is a single peak for each acquired waveform , then extract y array from the waveform cluster( you have already done it in ur VI) and connect it  to 'find array min max' function found in array functions palette and u get the single peak for each waveform.
    If it is a series of peaks u are interested in, then compare all these values inside a for loop.
    Index each element inside a 'for' loop, check it for threshold using a conditional case structure and put it inside a shift register.
    regards
    Dev

  • Peak detection/DAQ

    Hi! I am new in LabView.
    I try to solve a problem with a peak detection in signal. I use BNC-2110.
    This is my input signal - graph above show just digitized signal, and graph under - signal with "mean".
    I have found 2 suitable .VI. They found peaks, but in simulate signal. I have try to change this program for using DAQ-Date as input signal. But it was not successful, I have become many errors etc.
    Can you advise me, where is a problem?
    With best regards, Alex
    Attachments:
    Signal.PNG ‏78 KB
    Peaks_ermitteln2.vi ‏49 KB
    peakxmpl.llb ‏40 KB

    I now about this peak detectors, but they show very many peaks instead of 1
    And in this Program (peaks_ermittlen) it was very good realized. But I don't understand, how can I change the type of Input Signal.
    In this pic I have to draw, what i measure must - Time between 2 peaks
    With best regards, Alex
    Attachments:
    Unbenannt.PNG ‏48 KB

  • Fix for ORA-00060: Deadlock detected errors

    Hello APEX community,
    in the last days we are hitting more and more frequent the ORA-00060: Deadlock detected error in one of our applications. Logged an SR with Oracle Support but so far not too much help, except pointing to some older bugs:
    Bug 6618662: APEX APPEARS TO BE CAUSING DEADLOCK - CASE COLLECTION or Bug 7587013: INSERTING AND DELETING WWV_FLOW_DATA CAUSES DEADLOCK, which do not look at all appealing as the first bug was logged in Nov 2007 and the resolution was delayed from version 3.1, to 4.0 and even to 4.1, while de second was logged in Nov 2008 and not much happened with it.
    I am curious how many other users are hitting the same issue and which workarounds found for the problem. For example yesterday I could see 12 such errors in my log, today already 3.
    Florin

    Hi John,
    here it is: http://www.moyersoen.be/auction/1442/ or http://www.moyersoen.be/pls/apex/f?p=2008:11:0::::P11_AUCTION_ID:1442. Normally the page response time is < 0.20, but from time to time it goes up to more than 20 seconds and at those times I can see also the ORA-00060: Deadlock detected errors.
    Looks like the users are clicking twice refresh for that page. I am trying now to build a testcase that reproduce the issue easily.
    Florin

  • Calculating BPM using either Peak Detection point by point or FFT

    Hi Guys
    Im new to Labview and have absolutely no idea on programming and stuff. Im doing a project on Heart Rate monitor.
    I'm using labview to read the analog input to an Arduino Mini. In my attached VI im using Peak Detection Point by point to calculate the BPM but it doesnt seem to work. I took references from several VIs to arrive at my VI. 
    My instructor told me I could try using FFT to calculate the BPM as well but Im not sure how to carry it out in Labview.
    Hope you guys can help me with this.
    Thanks alot!
    Attachments:
    heart signal.jpg ‏43 KB
    Heart Monitor.vi ‏24 KB

    Ok, we have some problems here.
    1.  The Data Bits property is the number bits for a single character that is being transmitted.  You should not use that.  Since you are using an Arduino, it should be sending the termination character.  So just tell the VISA Read to read maximum number of bytes you expect from a single message or just some large number (like 25).  The VISA Read will stop reading the port when it encounters the termination character.
    2. The String Subset is not doing anything.  Just remove it.
    3.  You should move your  Wait to be outside of the case structure.  As is currently written, if you are not taking readings you will use up all of your CPU.
    4.  You should have labels for all of your controls and indicators.
    5.  Your time calculation is completely wrong.  You want to subtract the time of the previous peak from the current peak.  I recommend you use a Feedback Node to do this.
    Here's a slightly cleaned up version of your code.
    There are only two ways to tell somebody thanks: Kudos and Marked Solutions
    Unofficial Forum Rules and Guidelines
    Attachments:
    Heart Monitor_BD.png ‏42 KB

  • Peak Detection and Display of TDMS file of 4 channels

    The Peak Detection and Display sample VI in the NI Example finder is exactly what I am looking for but I would like to use a recoded TDMS file as its input. The TDMS file is a 4 channel data file.
    Does someone have an example code on how to do this?
    ERB 

    Hi
    I found an example in exampler finder called Read Channel Groups (TDMS).vi, which seems to plot data for four channels written in a TDMS file.
    Its also attached.
    Kind Regards, 
    Kind Regards
    James Hillman
    Applications Engineer 2008 to 2009 National Instruments UK & Ireland
    Loughborough University UK - 2006 to 2011
    Remember Kudos those who help!
    Attachments:
    Read Channel Groups (TDMS) forum.vi ‏20 KB

  • Peak detection in certain frequency

    Hello everyone.
    I am a rookie of labview, and I want to realize a program.
    The program reads in data and then do FFT transform.
    After getting FFT waveform, I want to search the waveform for peak in certain frequency, say 0.4~0.6Hz,
    and then output its magnitude.
    I have done the FFT part, but I have no idea about peak detection in certain frequency.
    Please help me! Thanks!

    Isnt this what you want ?
    Regards
    Asad Tirmizi
    Design Engineer
    Institute of Avionics and Aeronautics
    " Its never too late to be, what u want to be"
    Using LabVIEW 8.2
    Attachments:
    FFT min max.vi ‏111 KB

  • Preflight detected errors - overset text (1)

    Hi
    I have copied / pasted text from a Word document and tried 2 ways to place it in my InDesign CS6 document
    (1) putting copied text into a prepared text box
    (2) using the 'text box' that 'came with' the copied text (enlarged to fit area)
    each time the dreaded 'red box with a + ' sign inside appears on the rt. side of the 'finalised' text
    and 1 error shows at the bottom which is the Preflight detected error - overset text
    I have loads of other text in my document but none copied/pasted
    Writing the text manually into the document is not an option in this case
    Do I have to format it somehow? I would really appreciate a really simple A-Z step answer for this old gal
    cheers  - any help will really stop my hair from greying any further Lou

    Does the text contain special glyphs that require a font with Maori support? Id not then you can edit the styles to change the font (or, on the strong probability that the Word file doen't really use styles, you can use Find Font to replace it). if you do need Maori support, you should install a font that supports it, like Arial Maori, on the computer. ID uses any fonts installed normally for the OS (presuming they are not really badly made) or that are located in the private Adobe or InDesign fonts folders.

  • How to use peak detection in labview to detect peaks from data acquisition information

    Hi
    I am a university engineering student who is working in a team to develop a coin detector, its purpose is to recognize different coinage and detect fakes.
    For this we are using LabVIEW 8.5.1. I am relatively new to labview and have had no experience of using to before. We are trying to integrate four voltage signals produced by an electromagnet, straingauge, optical sensor and a proximity sensor. We have already developed signal conditioning for these tests and now wish to put them into labview.
    Our plan is to use peak detection on each of the tests so that labview can detect peaks which correspond to different coin types, provided they meet set criteria for each coin. Then to combine these either using logic or mathscript to produce a Boolean output for each coin.
    One of our advisors helped us develop a peak detection program for a simple simulated sine wave however we are struggling to adapt this for data acquisition information and itegrate it with mathscript and to be honest it does not make much sense. I have attached the program below. Thanks in advance for your assistance. 
    Attachments:
    Strain 2.vi ‏25 KB

    Chris,
    Here are several ways to help  you get started with peak detection:
    1. On your functions palette, you can search for "peak detect" and you'll find several different variations of VIs that will do peak detection.
    2. You may also want to take a look at this tutorial: Peak Detection Using LabVIEW and Measurement Studio
    3. There's an example in the example finder called "Peak detection and display" that will probably be useful. 
    Hope this helps, 
    Misha

  • [svn:bz-trunk] 21394: bug fix for watson 2887837 Not getting duplicate session detected error when same flex client id is used from two different HTTP sessions in CRX .

    Revision: 21394
    Revision: 21394
    Author:   [email protected]
    Date:     2011-06-16 12:34:13 -0700 (Thu, 16 Jun 2011)
    Log Message:
    bug fix for watson 2887837 Not getting duplicate session detected error when same flex client id is used from two different HTTP sessions in CRX.
    get the sessions id before we invalidate the duplicate session.
    Checkintests pass
    Modified Paths:
        blazeds/trunk/modules/core/src/flex/messaging/endpoints/BaseHTTPEndpoint.java

    For our profect I think this issue was caused as follows:
    Believing that remoting was full asynchronous we fired a 2 or 3 remote calls to the server at the same time ( within the same function ) - usually when the users goes to a new section of the app.
    This seemed to trigger the duplicate http session error since according to http://blogs.adobe.com/lin/2011/05/duplication-session-error.html  two remote calls arriving before a session is created will cause 2 sessions to be created.
    Our current solution ( too early to say it works ) is to daisy chain the multiple calls together .
    Also there seemed to be an issue where mobile apps that never quit ( thanks Apple! )  caused the error when activated after a few hours.
    I guess the session expires on the server and the error above occurs on activation.
    So the mobile apps now ping the server with a remote call when activated after sleeping for more than one hour.
    All duplicate http errors are silently caught and reported.
    Fingers crossed we won't get any more!

  • Detect error message on the screen

    How do I detect error message on the windows 2000 screen? I need to detect any error message appearing on the screen
    Thank you,
    Jeff

    You mean any kind of message generated by any kind of program, whether Java or not?

  • Looking for function or vi to calculate 1/2 power points of data obtained from peak detection vi

    I am analyzing vibration data and need a function or vi that will operate on peak data returned from the LabView Peak detection function and provides 1/2 power points or sideband frequencies. These are needed to compute a "Q" factor. So far I have not been able to find anything.
    Thanks, Fran

    The sound and vibration toolset may help here.  there are several vibration level measurements including peak.  With the advanced FFT functions in the Sound and Vibration tools, the zoom FFT can be used to identify and find sidebands and associated amplitudes. 
    here is a link to the manual
    http://digital.ni.com/manuals.nsf/websearch/1F8691E94728BDF5862570490058C3B2
    here is a link to the toolset
    http://sine.ni.com/nips/cds/view/p/lang/en/nid/10521
    please note the bullets on the above page (about half way down) noting peak detection, etc.
    Preston Johnson
    Principal Sales Engineer
    Condition Monitoring Systems
    Vibration Analyst III - www.vibinst.org, www.mobiusinstitute.com
    National Instruments