Hello. I am looking for help increasing the execution speed of the attached VIs. The equipment is a PXIe-8135 controller running LVRT, PXIe-6368 DAQ, and PXIe-7966R FPGA. I am using a hardware timed, single point data acquisition loop that will eventually implement some sort of feedback control, and I need the highest possible rate (ideally above 100 kHz). Leaving out the FPGA, I was able to run the loop at 200 kHz on the real time PXI target. However, incorporating the calls to the FPGA has cut that down to about 60-70 kHz. (The FPGA is needed because multiple channels will be added and the computation might get more involved than the RT processor can handle quickly.) Here are some specific questions.
1. Data conversion. Now I am reading data from the DAQ as a floating point double, but then I need to convert it to send data to the FPGA. I am not sure if converting to fixed point rather than single is more efficient? Or should I try reading raw I16 data from the DAQ? I believe this creates an array of integers, so it might be slower?
2. Data logging. If I turn on data logging, the rate is drastically reduced. Is using the built-in TDMS logging from the DAQmx pallete not the best option? This usually works better for me than a parallel data logging loop with local variables.
3. FPGA polling. Would another method be better? For single point transfers, I didn't think DMA FIFOs are the right approach.  I did try the IRQ interrupt method, and it was slightly slower.
I really appreciate any suggestions. Any little tweak will help at this point. Thanks for your help!
Jeff
Attachments:
fpgaVI.png ‏11 KB
targetVI.png ‏25 KB

You are going the expensive route!  Are you using any of the IO on that FlexRIO board?  If not, you are totally going in the wrong direction with this.
Why not just go with something like the PXI-7854R, which can handle your analog inputs and outputs.  And it is cheaper than that DAQ card.  Then all of your acquisition and control logic can happen on the FPGA.  You just have to use a DMA to send the data to the RT system for logging.
There are only two ways to tell somebody thanks: Kudos and Marked Solutions
Unofficial Forum Rules and Guidelines

Similar Messages

  • Optimum servo pid loop rate

    Hi,
    I want some confirmation on the optimum pid loop rate and the fastest servo motor operation that we can acheive with Pxi-7352 running on Window.
    The manual of 7352 says "62 μs PID loop update rate for up to 2 axes". The requirment is to acheive periodic equi sized straight line and rotary motion ( using yaskawa servo motors in torque mode and 7352 to close position loop ) Both motion are independant and not operating simultaneous i.e rotary completes one full rotation in small steps and then linear moves by one small step then rotary completes next rotation in small steps and again linear motor moves this continues till end of test.
    you can observe here that both rotary and liner motion are periodic and move equi spaced distance for each step with small stoppage/dwell time of 500 u sec , The catch is in addition to positional accuracy in terms of 50 -100 microns the time for complete each step is also very crucial and very small (many peridoc steps together consitute complete linear and rotary motion). Find expected velocity profile attached for more details on desired timing.
    As the motors have simple and periodic straight line and rotary motion i suppose buffered contour moves are not needed which have limitation of 10 msec between two points, then can the time to acheive each step be as low as 2.5 to 5 msec considering best loop rate of 62.5 u sec and highest/worst pid loop rate of 500 usec.
    1. Is the expected timing of 2.5 - 5 m sec feasible assuming that the motor is capable of reaching 5000 rpm which is well above desired speed.
    2. 735x series being high performance motion controller what is the order of closed loop timing that we can expect for simple and periodic moves (say in order of 1000 usec or 10 msec)
    3. Can i interface one stepper motor and another servo motor where both the motors and drives are in position mode on single 7352 and UMI 7772.
    4. Are there any limitation if NI motion is not used for code development and NI motion Assistant along with LabVIEW is used instead.
    This is precision application and the component is very small ,the weight of the component would be less than 400 grams so i hope we do not have to consider inertial affect on motion significantly.
    Appriciate if some one with similar experience can provide more details.
    Regards,
    Bipin
    Solved!
    Go to Solution.
    Attachments:
    Velocity Profile.png ‏21 KB

    Hi Bipin,
    thank you for your comprehensive post. I had a close look at your timing specs and I strongly doubt, that these specs can be met with a 7352 board. Even though the PID loop rate of 62.5µs might be enough, both trajectory generator and the supervisory control (host communication, move command sequencing,...) are too slow to meet your requirements. Please note, that the trajectory generator always works with a timing interval of 10 ms, regardless of the operation mode (linear move, contouring, blending,...), so there is no way to reach your timing specs with this product. In general contouring is the fastest way to run a quick succession of moves, as it doesn't require host communication (by the way: onboard programs don't execute much faster, than programs that are running on the host).
    I don't want to make negative mareketing for our 735x boards, but your timing requirements are just way beyond the requirements of typical motion control applications and it's probably quite hard to find any COTS product in the market, that can meet these specs. 
    With this said, I would like to offer you an alternative solution, that is based on NI's FPGA technology. Additionally to a faster trajectory generator a faster control loop is probably also very desireable for your application. Even with the lowest price Virtex-II based R-Series multifunction board (PCI-7830R) you can run PID loops with 5µs cycle time and the trajectory generator at the same rate. You even can use a different type of control algorithm if required. This technology is leveraged very successfull in a lot of applications with very high performance and accuracy requirements (here are some examples).
    So these are the pros, but where are the cons? Well, the R-Series boards are general purpose FPGA boards and not ready to run motion control boards, so you need to develop the low-level control application by yourself. The good news here is, that it can all be done graphically in LabVIEW and that there are many ready to run examples for motion control, that can be used as a base for your development. Additionally you may want to have a look at the NI SoftMotion Development Module for LabVIEW, that simplifies the development of motion control applications, but it is not a mandatory tool and the benefits of this module might not be as high for your application, as ithey are for more standard motion control applications.
    I hope that helps.
    Kind regards,
    Jochen Klier
    National Instruments

  • [FPGA] Loop rate very slow: Do FPGA I/O nodes in parallel loops block each other?

    Hi,
    I am using cRIO-9075. Mod1 is NI 9263, Mod2 is NI 9227, Mod3 is 9215.
    Please see my VI attached or the given screenshot.
    The FPGA code is based on the "NI CompactRIO Waveform Reference Library" (it's the lower loop).
    The upper loop was added by me and is writing a waveform from blockmemory to the NI 9263 module (Mod 1).
    The data sampled in the lower loop is running at 1 kHz. The control "AO Update Period" for the upper loop has a value of (for example) 10 (=uS).
    The problem is, that this loop is running much much slower than it should. Once I disable the FPGA I/O node in the lower loop (as done in the attachments), it's running as fast as it should.
    It seems to me, that the FPGA I/O nodes are blocking each other. I tried to figure it out by reading through serveral NI documents, but until now I have no idea how to solve that.
    Can you give me some advices? Some general tipps about the VI?
    Thanks!
    Attachments:
    FPGA Loop Rate.PNG ‏72 KB
    FPGA Main.vi ‏251 KB

    Hi, thanks so far.
    Originally the control was inside the loop. Then I tried if it makes a difference if it's outside.
    Ok, i really seems to be that default value of "100000" for "AO Update Period".
    Starting the VI directly woks like expected. Having "AO Update period" inside the loop makes it possible to control it as it's running.
    But, please see the attachment. When starting the FPGA through RT and setting the appropiate value, it does not seem to work. The oscilloscope show's the same behavior like "AO Update Period" was 100000.
    But when reading the value of "AO Update Period" afterwards (while the FPGA is running), it shows the expected value of "10".
    Having changed the default value to 10 works so far, but I am not able to changed it (see attachment).
    So the problem is: Why is "Read/Write control" not working here? Why is still the default value used?
    Attachments:
    FPGA Loop Rate 2.PNG ‏5 KB

  • Need to slow down POLL loop rate

    I find that when I configure a device for polling and a connection is opened, the poll rate appears to run as fast as the computer will allow. I need to control the poll loop rate or "scan rate" so it scans once every 100ms. Some devices cannot handle a fast scan rate, must be slower than 10ms. How do I change my scan rate?

    Hello,
    Thanks for contacting National Instruments.
    I think you can accomplish what you are trying to do by using a different mode instead of Automatic. For example, if you use the Scanned mode, then you should be able to set the ExpPacketRate to the value you are interested in using as the poll rate. This information is documented on page 2-59 of the following manual.
    http://digital.ni.com/manuals.nsf/webAdvsearch/659F311DE89061978625680F0059DCF7?OpenDocument&vid=niwc&node=132100_US
    Hope this helps!
    Regards,
    Steven B.
    Applications Engineering
    National Instruments

  • PID loop rate with counters

    I'm trying to use a PXI-8106 RT and LabviewRT 8.5 as a PID controller to control 3 motors.  I'm measuring the speed of the motors by measuring the frequency of the encoder with a 6608.
    This article http://zone.ni.com/devzone/cda/tut/p/id/5423 claims a speed up to 107.5 khz.  I've set ethernet to polling and usb off, I didn't turn off  the CPU display, but I'll try that as soon as I can. However, I'm not anywhere close to 107khz, best I can get is about 10khz with one channel and about 2 khz with 3 channels.  I realize the tests were done with analog inputs, but I was hoping that  I'd get at least somewhere close using counters. With the 2Khz rate I really can't do what I need to do.
    Does anyone know what a realistic rate would be with this setup?
    I'm new at Labview so I may be doing something wrong, although I can't see what it could be. I've done quite a bit of research on the forums before I started to learn from other's examples.
    Here's a couple of things I've learned  trying to optimize my loops.
    Seems like Daqmx has some issues with applying a scale.  I have 2 scales, one is just a scale factor, 0.166, the other is .0002.  When applying the .0002 factor in Daqmx my performance takes  a huge hit, something like 10%!  If I just leave it at hz, and do the scaling myself I see no performance penalty.
    There is something strange about the timed loop % display in RT8.5, I wrote my own routine using expected end, actual end time, and the period to give me a % loop useage and it matches very well to the RT8.5 display for most of the time, however, when the loop starts missing iterations because I've gone over 100% the RT display drops down to about 50% which pretty much makes it useless.
    I've attached a couple of shots of my test loop. I can get this one to almost hit 10 khz. the other screen shots are of how I set up the counter, and how I read the counter. the read counter has the error trap to restart itself.  Also I set it to use DMA explicitly somewhere else in the program.
    Thanks for any help.
    Henri
    Attachments:
    Main Loop.JPG ‏55 KB
    Setup Counter.JPG ‏72 KB
    Read Counter.JPG ‏67 KB

    Devin,
    Thanks for the suggestions.
    A couple of related questions for you.
    The 200 uSec is leftover from my attempts at getting the 3 PID loops to work within the loop availability.  I can run  about 70 uSec i.e about 14khz  with only  a single counter.
    I've tried changing the buffers value and it has barely any effect on the speed of the loop. Right now I have 3 counters running with 200uSec timing and I get about 69% loop utilization. With a few missed iterations every once in a while. This is using 10000 samples, if I drop down to 100 samples utilization actually goes up to 70%!?  Doesn't make sense to me either, but this is using both my VI and the on screen Timed Structures % of LV8.5RT so I'm trusting it as real,
    In theory if I can get my loop rate up to 10 khz I'll only need 10 for the buffers since my top frequency to be measured is 100 kHz.
    Funny thing on the RT fifo variable... I turned on FIFO and it actually decreased performance by about 1%, both Total load and Time structures went up... so I turned it off for stop-RT, and got a 3% improvement! Down to 66% time structures.
    You mentioned the hardware timing to get a hardware timed single point. Unless I misunderstood the counters, this isn't possible since the sample rate is implicit in the frequency measured? i.e. it latches the count every rising edge.. so I'm my case I'd get a 100 khz sample rate (although obviouslyl that would be the max) Can you let me know if I've missed something there.
    I bought this system explicitly based on its dual core and its high speed so I could close my PID loops quickly enough with one CPU and still have processing time left over to log data, etc.  And the 6608 card is there so I can get high accuracy for my frequency measurements.   However, I'm really struggling to get this thing going.
    Any help would be greatly appreciated.
    Thanks
    Henri

  • Reading Data from a Queue - Store in Fast Loop, Display in Slow Loop

    Here's my BD below.  I am queueing data in a faster loop (either 100 ms or 50 ms) and sending that data to a 250 ms loop.  I need to be able to display the data in the slower loop (so that the FP doesn't have to update every 50 ms).
    Problem I'm having:  How do I empty the data?  It builds and builds and eventually I'm looking at a chart that has data that's 1 minute old.  I've tried writing my own flush code (attached).  I've tried to use variables to pass data and use an array to parse...not working.  Any help would be appreciated.
    Nathan - Certified LabVIEW Developer
    Solved!
    Go to Solution.
    Attachments:
    FlushQueuebyNathan.vi ‏21 KB

    If you need all of the data to go into the chart, I would use the flush queue and write to the chart in a FOR loop (autoindexing with the elements array from the flush queue).  You may set the Defer Front Panel Updates before the FOR loop and reset it after, but I'm not sure that will buy you that much.
    There are only two ways to tell somebody thanks: Kudos and Marked Solutions
    Unofficial Forum Rules and Guidelines

  • PCI-7041 pid loop rate

    Hello,
    I have PCI-7041/6040 card with which I would like to perform single point PID control. I have looked at the DAQmx example PID control - single channel.vi to do this. The only thing I am unsure on how to do, is set the loop rate to a specific kHz value. Is this done by using the 'Input Rate' control and how could I verify that the loop is running deterministically and to the rate I have set?
    thanks,
    David

    Hi David
    You are correct that the "Input Rate" control will set the loop rate for the example that you are using. There are a few ways to determine if the loop is running deterministically.
    One thing you can do is use the "DAQmx is Read or Write Late" VI. If the "is late?" output goes true, your loop is not keeping up with the specified hardware timed acquisition and generation.
    Another recommended method to determine if your loop is running determinisitically is to benchmark it using the TimeStamp VI's.
    There is a great example in the LabVIEW Example Finder that you can use for this. From the Example Finder, go to
    Tookits and Modules >> Real-Time >> Timing >> NI Timestamp Code Timer.
    You can combine your PID example code with this example to determine how long it takes for your PID code to execute when running on the real-time target. Just place the code you wish to benchmark in sequence 0 of the sequence structure of the Timestamp Code Timer example VI. This VI will give you some great information including average execution time and maximum jitter.

  • FPGA - Synchroniz​ation between two loops with different loop rates.

    Hi all,
    I am trying to implement an Analog and Quadrature Decoder. Simply put, I have two loops:  LOOP 1 with an Analog Input (AI) which is a slower loop rate. Another loop: LOOP 2 has a digital input (input) and counts the Quadrature counts. I understand that Analog conversion is slower, and as a result,  LOOP 1 will have a slower higher tick rate. I want to pass the "counts" from LOOP 2 into LOOP1, so that I can synchronize the Analog Input with the number of quadrature counts. My issue at hand is: since LOOP 2 updates at a higher rate. Will the "counts" value that I pass/tunnel from LOOP 2 to LOOP 1 refresh at the LOOP 2's rate or at LOOP 1's rate.
    Thanks,
    Herrick Chang

    hlchang wrote:
    Hi all,
    I am trying to implement an Analog and Quadrature Decoder. Simply put, I have two loops:  LOOP 1 with an Analog Input (AI) which is a slower loop rate. Another loop: LOOP 2 has a digital input (input) and counts the Quadrature counts. I understand that Analog conversion is slower, and as a result,  LOOP 1 will have a slower higher tick rate. I want to pass the "counts" from LOOP 2 into LOOP1, so that I can synchronize the Analog Input with the number of quadrature counts. My issue at hand is: since LOOP 2 updates at a higher rate. Will the "counts" value that I pass/tunnel from LOOP 2 to LOOP 1 refresh at the LOOP 2's rate or at LOOP 1's rate.
    Thanks,
    Herrick Chang
    The only way to pass values between two parallel loops is by some sort of global variable (global or locals are fine, as are shared variables although I wouldn't recommend that for FPGA if you don't want to go outside of the FPGA target at all). Don't tunnel anything between two independant loops as they won't be independant anymore.
    The local variable will be updated at the rate the writing loop updates it, though your slower read loop will obviously not see every value. Note that the FPGA compiler will protect the local variable access in both loops in order to avoid race conditions. This protection is necessary but expensive in terms of runtime performance of each loop as well as FPGA gate usage.
    Rolf Kalbermatter
    Rolf Kalbermatter
    CIT Engineering Netherlands
    a division of Test & Measurement Solutions

  • Ni7842 FPGA RIO board loop rate

    Hi,
    Is it possible to achieve a loop rate of 1us with the Ni7842 RIO board. It has an onboard clock of 40MHz. I have tried the following approaches to achieve a 1us loop rate and generate a single pulse of 1us duration, but wasn't successful.
    1st approach: The above flat sequence was originally with 2 frames, I removed the second processing frame to reduce any possible processing delays and changed counter units to usec with a count of 1. This did not work, minimum pulse width I could generate was 1ms.
    2nd approach: I changed counter units to ticks with a count of 40. This also was able to generate a min pulse width of 1ms only.
    The AO channel is capable of 1MS/s update rate and with an onboard clock of 40Mhz you would think you can achieve 1us. Why is it being so difficult for me? Please help.
    Thanks,
    PG

    Show some actual code.  If possible, supply a stripped down project that shows this issue.  I recommend zipping up this project and any needed VIs and posting the zip file.
    There are only two ways to tell somebody thanks: Kudos and Marked Solutions
    Unofficial Forum Rules and Guidelines

  • My nano plays back itune podcasts at too fast a rate.

    Hi All!  My nano plays back itune podcasts at too fast a rate.  My iMac does not do this with the same podcasts.  If I have my nano connected to my iMac and use it (the nano) to playback thru my computer monitors- it's fine.  Unplug it and use earbuds and it's too rapid!?  Did my nano drink too much coffee?  Thanks, Michael

    O.K.  Finally figured it out...  Podcasts have a 1/2x, 1x and 2x speed setting on one of the windows that will change the playback speed.  Never Mind!

  • Maximum loop rate achievable in FP 2010

    Hi..
    I am developing and application with FP-2010 RT controller. Right now I am in design stag and wondering what will be the maximum loop rate achivable with this controller.
    searched ni.com... but didnt get any clue..
    I hope some of you would have experienced with this controller.
    Any help towards this will be greatly appreciated.
    MIRASH

    Hi Mirash,
    just now I try also to improve the performance of a control loop with the FP-2010.
    Up to now I find a limit of 100 msec to transfer one item.
    My next trial will be to combine some items to one transfer.
    Additionally I ask the question NI but have still no answer.
     skyh

  • Loop rate on my RT system not faster than 1kHz

    Hi,
    I have a PXI5404 frequency generator card without memory. So I tried to
    program the memory list on my own with  while-loop. The value for
    the frequency card is updated with an extern digital signal. But it
    sems that my loop doesn't run faster thn 1ms. Is there any solution?
     I have a PXI system with an PXI 8176 Controller, an 6259 M-series card, a PXI 6723 analog out card and the frequency card.
    Thanks for your help
    Tim
    Attachments:
    Loop.vi ‏90 KB

    Thanks for the answer.
    In deed, I was able to change my frequency faster, down to 200 microseconds. First I gave out a frequency of 20kHz, switched to 3 MHz for 200 µs and back to 20 kHz. So, the card can read the frequency faster, but after I switched back to 20 kHz, the frequency isn't updated any more. The card is missing one trigger. Could it be faster to communicate directly with the analog devices chip un the card via some hardware addresses?
    Thanks
    Tim

  • How to combine Digital Output, a delay and Analog Input in a fast loop

    I need to develop a process loop that runs at least at 250 Hz that performs a Digital output, than a delay of 50 microseconds and than an analog input of all the channels. All will be done using ATI MIO 64E3 card. Of course, the acquired data will be processed, displayed and saved. The loop will be running for several minutes until user stops it.

    The fastest and most precise timing will occur if you use hardware timing. You can apply hardware timing to analog input on the E Series boards, but not the digital lines. Let's focus on the analog input first. Continuous waveform scanning uses a scan clock, which can be the board's internal one or an external one which you apply. If you want to scan all the channels 50 microseconds after a digital rising edge, then you need an external signal to signify that scan clock.
    The E Series boards also have 2 counter/timers onboard that you can use for this purpose. You can set up a retriggerable pulse generation operation, where the counter receives a trigger and then on the user specifications, produces a pulse. You can have that route to the analog input scan clock.
    The trigger signal for the counter is that digital pulse. As I mentioned earlier, there is no hardware timing for the digital lines on an E Series board. We do have other digital boards (653x family) that have hardware timed operations if precision is important. If you are satisfied with software's resolution (in the milliseconds), then you can call the E Series board digital function in a loop with a software timer. That digital line can route to the counter to act as the trigger.
    So, on the programming side, you can have three separate and independent operations in parallel. One is for the digital function to output on that line every so often. Another is for the counter set at the retriggerable pulse generation. The last is for the analog input. I will describe this in terms of LabVIEW, but it can be done in a similar fashion with the NI-DAQ function calls or Measurement Studio.
    The digital examples are in the LabVIEW >> Examples >> Daq >> Digital >> E-Series directory. The Generate Retriggerable Pulse example is in the LabVIEW >> Examples >> Daq >> Counters >> DAQ-STC directory. The E Series boards use the DAQ-STC timing chip.
    Go to the LabVIEW >> Examples >> Daq >> anlogin >> strmdisk.llb directory and start with the Cont Acq to Spreadsheet File. This shows how to continuously acquire data and stream it to disk while displaying the data on a chart. Streaming to disk is the efficient way to save data while you are acquiring, as it eliminates the overhead of always opening and closing the file through the iterations of the loop. This saves to a file that can be opened by other applications (Excel, Word, etc.), but it is not as fast as writing to a binary file, which must be opened and read back through LabVIEW. However, for your ~250 Hz rate, it should be fine. Then, go to the LabVIEW >> Examples >> Daq >> anlogin >> anlogin.llb and look at the Acquire N Scans -ExtScanClk example. This shows how to apply the scan clock. Here, the AI Start that you saw in the previous example is replaced by 4 VIs (3 AI Clock Config's and the AI Control). Make those changes to the first example and then add a constant 0 to the AI Control parameter for total scans to acquire. That specifies the continuous operation. The File >> VI Properties >> Documentation menu item of the example describes the physical connections.
    If you aren't using LabVIEW, use the NI-DAQ User Manual and the NI-DAQ Help file installed on your machine. You can look at your AT E Series User Manual at the http://www.ni.com/manuals pages for more information on the hardware. Also, if you want to route those signals internally on the board, you can find some entries in the KnowledgeBase at the http://www.ni.com/support pages.
    Regards,
    Geneva L.
    Applications Engineering
    National Instruments
    http://www.ni.com/ask

  • Simultaneous save of a fast loop and a slow loop with multiple mx inputs.

    I am attaching my app in hopes that someone might help me. In the app I have two loops. One takes in data from 2 inputs (temp and voltage and I would like to add one more voltage on another channel), the other runs very slowly through the comm port to a mass spectrometer. I can not find a way to graph all the data in one simple file and just as annoyingly is the lame attempt to read temperature in both loops. This method ensures that my Mass spec slow loop data will have its own temperature readings but it errors sometimes do to the way I have the MX input. I tried doing all the saving using queues but the slower mass spec loop was not offering up data quickly enough causing the dequeue loop to wait for data and then the upper quick loop queue to pile up. So i have two issues/options -#1 figure out how to save data from both loops to a single file, and fix the temp inputs so they wont crash each other. Or #2 use the queue method but not have to wait for data from the lower slow loop. Any ideas? Really what I am trying to do is not complicated, the slow acq rate of the ms kinda ruins things. It spits out data whenever it feels.
    THanks
    Attachments:
    TPDS1.vi ‏697 KB

    Hello Tim,
    A producer-Consumer solution using queues should solve your problem. I have attached below a starter for what I am refering to. There are two producer loops, each of wich will enqueue the data and a tag for the data. The data and tag are clusterred so that you can check the tag whenever you dequeue the data.
    In the consumer loop, you can check the tag every time that you dequeue an element and act according to the tag. You can even dequeue directly to file and save both the data and the tag, that way when you read the file, you can check the tag and act accordingly.
    Hope it helps.
    Attachments:
    Acquisition.zip ‏36 KB

  • Loop rate long

    Hi,
    I have a PXI5404 frequency generator card without memory. So I tried to
    program the memory list on my own with  while-loop. The value for
    the frequency card is updated with an extern digital signal. But it
    sems that my loop doesn't run faster thn 1ms, altthough it is an RT system. Is there any solution?
     I have a PXI system with an PXI 8176RT Controller, an 6259
    M-series card, a PXI 6723 analog out card and the frequency card.
    Thanks for your help
    Tim
    Attachments:
    Loop.vi ‏90 KB

    duplicate post

Maybe you are looking for

  • Moving images from iPhoto 11 to Photoshop Elements Organizer

    I understand that iPhoto 11 kept all images within the iPhoto Library.  In order to move them into the Adobe Photoshop Elements Organizer, they must be moved to the hard drive of the Mac and then imported into Elements 12 Organizer. Is there any way

  • How to find out Export area for a import

    Hi All, I have a issue ..in my program we are inporting one value from ABAP memory . iwant to know where from it is exported i try with debugg by seeing system merory it's showing the filed name but it is not showing where from it is exported pls hel

  • Content of the audigy 2

    Ok i have just bought this http://www.ictcompany.com/store/prod...79&Item=SB0240 can someone give me a list of all the stuff that is on the cd that is supose to come with the card, so that i can make my own cd.

  • Canon HF20 won't mount after upgrade to Snow Leopard

    Since upgrading to Snow Leopard, my Canon HF20 will no longer connect via USB. This is a basically a USB hard drive as far as the OS is concerned, and I've typically mounted the camera and copied my AVCHD files to an external drive for editing. No sp

  • Can I use older actions in the CS6 Beta? Haven't tried yet--but want to

    I am loving CS6 Beta, and I hope to be able to load my favorite older actions without hassle. Any advice?