I use the PXI 7344 to control a wave paddle in our wave flume.
We make regular and irregular waves.
At this moment I use a timed loop to update the position of the paddle 40 times per second (loop time 25 ms).
I use a linear interpolation between the current and the next position.
The reason for not using the contoured move is as follows:
We have also a active wave absorption system programmed into our controller. It uses 2 analog input of the PXI 7344 and calculates a correction/feedback for the wave paddle.
Therefor the movement of the paddle is not predetermined and cannot be predicted.
I would like to use the contoured move to generate a smoother motion of the paddle.
Is there a way to correct the motion during the exection of a contoured move (like the analog feedback) but then programmatically (by a calculated value ) ?
Thanks,
Tom
Tom Versluys - Dept. of Civil Engineering, Gent University
Technologiepark 904, B-9052 Zwijnaarde (GENT), BELGIUM
TEL.: +32(0)9/264.54.87, FAX: +32(0)9/264.58.37, MAIL:[email protected]

It may be that you have already chosen to map the 7344 to the local machine. That was the case for me - I have a PXI 7344 and PXI 7350, where the 7344 was already mapped to my local machine, and the option to map to local machine was grayed out. Try pressing F5 to refresh MAX and then check under the main Devices & Interfaces tab for the PXI-7344 on IP address.
Hope this helps.
-Julie

Similar Messages

  • PXI 7344 motion controller as a DAQ device for ELVIS

    Hi,
    Is there a way to reconfigure a PXI-7344 motion controller with its two 68-pin connectors so as to use it as a DAQ device that would receive the signals from a NI ELVIS workstation. The 7344 has a "motion I/O connector" and a "digital I/O connector". Analog outputs are even possible if I remember well, as well as digital.
    Thanks

    Hello,
    The card does has some analog input and outputs. I would check the manual here: http://www.ni.com/pdf/manuals/370838b.pdf. The DAQmx API with a DAQ card is going to be much better than interfacing with a motion card, but you should be able to read off it at the least.
    Rahul B.
    Applications Engineer
    National Instruments

  • Is it possible to condition a load cell signal with a 1520 SCXI and then use it as the primary analog feedback for a 735X motion control board?

    My system
    PXI 1052 chassis
    PXI 8196 RT controller
    Flexmotion 7356 motion controller
    Flexmotion 7354 motion controller
    6259 M series DAQ
    1520 SCXI
    1520 SCXI
    1314 SCXI frount mounting terminal block
    DSM strain gage load cells
    Exlar SR Brushless PM motors
    Kollmorgen servo star CD series 5 drives
    My objective:
    I want to measure the force on the load cells with the 1520 SCXI (i.e. excite the load cell then amplify and condition its output) then I want to use this analog (force) signal as the primary feedback for my motion control system (7356 and 7354 motion controller connected to my drives and motors).
    The problem:
    The motion controller boards (7356 and 7354) can be configured for force feedback control via analog feedback from a load cell.  However the boards implicitly expect the load cell signal to be amplified and conditioned with and external signal conditioner rather then using an NI SCXI module.  Currently I’m am reading the SCXI signals via the 6259 M seriers DAQ and consequentially have no way to route the signal into the 735X boards for force feedback control.  I can use softmotion to control the motors but then I loose the fast PID update rates possible with the 735X boards.
    The question:
    Is it possible to condition a load cell signal with a 1520 SCXI and then use it as the primary analog feedback for a 735X motion control board? 
    Thanks
    Patrick Aubin
    Ph.D Candidate
    University of Washington &
    VA MedicalCenter of Excellence for
    Limb Loss Prevention and Prosthetic Engineering  

    Hi GG,
    Thank you for your feedback.
    >create your own external circuitry to amplify the signal.>
    True a few op-amps could amplify and filter the signal externally but the performance of such an amateur built system will never match the accuracy and functionality of the 1520 SCXI.  That’s why we spent the $2500 for the professional version.
    >One thing that you could do is to use a UMI with your motion controller and then wire your signal from your SCXI to the UMI.>
    As I understand it the UMI is basically a break out connectivity interface for third party drives.  After conditioning and amplifying the analog force signal with the SCXI module the analog signal isn’t available to wire into a UMI.  To my knowledge the conditioned signal of the SCXI resides only in the PXI backplane (i.e. there is no physical access to the conditioned analog signal).  Can you explain how the UMI could be help here.
    Again thanks for your input GG.
    Patrick

  • NI Motion + PXI 7344 + s-curves

    I use Labview 7.0 and a PXI 7344 motion board.
    I try to calculate the time and the distance needed for a motorized stage to
    reach a constant speed.
    The Speed, Acceleration / Deceleration and S-curves parameters are known.
    Is there any way (vi or dll) to obtain the theoritical position, speed and
    acceleration Vs time of a movmement ?
    Or is there a way to access the data of the trajectory generator?
    The "motion assistant" plot the position, speed and acceleration Vs time of
    a movmement from the Speed, Acceleration / Deceleration and S-curves
    parameters. Is it possible to have the formulas ?
    Thanks for your help.
    Stéphane.

    Stephane,
    You can use the motion assistant to visually determine the response of the parameters that are entered. At this time the exact formula is not available.
    Austin Talley
    Applications Engineering
    National Instruments

  • What I can do to minimise my offset in velocity in PXI 7344?

    Hi,
    I am using Motion control Card (PXI 7344) to control both position and linear velocity of the cylinder shaft. I am reading analog voltage from potentiometer(input range = -5V to +5V corresponding to positions 100mm and 0mm respectively). I have used DAC output (+10 v to -10V) of the motion control card as in input to Servo valve (Note: i have not used any amplifiers inbetween) which controls the hydraulic flow which inturns controls the linear movement of the cylinder shaft. In the MAX i have used Axis 1 and configured it to velocity mode selecting only the primary feedback and primary output (I haven't used secondary one) and set PID update period to 62.5 microseconds.
    The problems I am facting righ
    t now are
    1. In the 1D interactive control of the MAX I found the current trajectory data (position and velocity) keeps on changing before starting the test (when the station if OFF) after intialising the card. I have check the Analog feedback signal from the potentiometer. It remained constant without fluctuating.
    2. I can't able to achieve the set point velocity (i.e., there is always an offset and it varies with increasing setpoint). So I have used Auto-tuning option in the MAX. It automatically sets some PID parameters but it remains changing when i do it again and again.Eventhough i make changes in the PID parameters, my offset remains the same.
    Whether I have to use Encoders as secondary feedback device or else what I can do to minimise my offset in velocity?

    Project Support,
    After reading over your description, I am still not sure that I entirely understand what is happening. Some of the questions that I have are:
    -What type of motor is hooked to Axis 1?
    -What type of feedback is Axis 1 using?
    -Does the linear movement have feedback?
    If you post back the answers to some of those questions, this forum may be a little bit better equipped to help answer your questions.
    As for the questions that you posed, if you are using analog feedback, and the system issuing that feedback is turned off, then you may be seeing fluctuations from noise in the environment. This is my best guess about what is happening with problem #1. As for problem #2, I don't fully understand what you are asking. If you could post some a
    dditional details about this problem, it may help to resolve the problem.
    I wish you the best of luck with your application.
    Regards,
    Jed R.
    Applications Engineer
    National Instruments

  • Using PXI 7352 Motion card and 6025E DAQ

    Dear Sir,
    I am using SCXI 1520 along with PXI 1011 & E Series DAQ (6025E) to measure the Load Value of a load cell. I am also using PXI 7352 Motion card to control the velocity of a Plunger. Now my requirement is to stop the movement of the Plunger depending on the value I read from the load cell. I am not able to achieve the stoping accuracy for high speeds (around 150mm/sec) in the software. Can it be possible with the use of Motion Control card and RTSI signals? If so, how can I do that? (I am not used any limit switches as the load value at which the Plunger should stop is user programmable and I am using DAQ card in Multiplexed mode because i am reading more number of analog inputs with only one DAQ card)
    Ramkumar. D
    Soliton Automation,
    India.

    Hello.
    Based on your post I am guessing that you are performing the control in software, i.e. starting a velocity move, and then reading the load cell until it reaches a threshold, then stopping the motion. As such you are essentially limited by the software control loop rates.
    You are using PXI so it is possible that you are using real-time. If not, one solution could be to implement a real-time application that would be more deterministic, but again this would still depend on the software.
    Depending on your system (stepper or servo), you could implement a direct analog feedback into the motion card itself, but this would require a signal which has a range much greater than your load cells. This is only a good option for servo axes. If this was somehow possible, you can tell the motion controller to move to a specified LOAD instead of a specified position/velocity. Provided you could appropriately condition the signal, this is the best choice.
    Unfortunately, you cannot implement limit switches/etc over RTSI, so I believe that this will not be a good option.
    A third, and probably most un-desireable option, is to output a voltage on the PXI-6025E proportional to the load received from the load cell. This voltage could then be used as the analog feedback for your motion device. This, however, still adds the software layer of delays.
    I hope these ideas help you out.
    Regards,
    Colin C.
    Applications Engineering
    Colin Christofferson
    Community Web Marketing
    Blog

  • HSC limitation​s of the PXI-7344

    I have an application where I would like to measure position data (encoder and analog value) from a relatively fast moving mechanism - so want to use the highest sampling rate possible while keeping both measurements synced. I am using the PXI-7344 for motion control and the PXI-6221/PXIe-6361 for analog DAQ.
    My approach is to use one RTSI line to provide the DAQ sampling clock (possibly divided with a counter for even higher sampling rates with the DAQ card) to the HSC trigger of the PXI-7344. According to the PXI-7344 datasheet, the maximum trigger repetition rate is 1kHz, but the HSC has to be re-enabled in software after every capture. I have a program doing exactly that and it seems to be working fine - but I can't achieve rates on the HSC higher than approx. 150Hz. I assume the bottleneck here is either the host communication (most likely) or the PXI CPU is too slow with re-enabling the HSC (very unlikely - it is a 2.7GHz Quadcore and I am polling the HSC status as fast as possible in a while loop w/o any waits).
    My next approach would be to write an onboard program to do the same thing and write the positions into an onboard buffer which I would read after the program is done, but according to the NI-Motion help these programs run inside the RTOS on the onboard µC with a repetition rate of 2ms, which would mean a maximum HSC rate of 500Hz (theoretical).
    This is of course better than 150Hz, but now I am wondering how I could use the full capabilities of the trigger input of the card (e.g. achieve a HSC rate of 1kHz). Is this even possible/foreseen with the 7344?

    HSC rearming rate is system dependent and affected by many factors. 100 Hz is a rate, that can be achieved on most systems. You might higher rates than that, but that's not guaranteed.
    Onboard programming typically doesn't result in much better results, as the onboard CPU runs the user programs in a low priority task, which gets interrupted frequently by high priority tasks like limit switch monitoring and so on.
    In general that's the bottleneck, that explains the limitation. On a 7344 HSC needs to get rearmed after every trigger occurrence in the user program and in both cases (host and onboard programming) the rearming needs to be executed by a low priority process on the onboard CPU.
    By the way: You are referring to a 2 ms interval. In fact, that's the default value of a time slice assigned to an onboard program. This is not equal with the onboard program's cycle time, which is typically higher - depending on the amount of code and the interrupts caused by higher priority processes.
    I hope that explains the behavior. For higher HSC rates up to 2 kHz please use a 7350 board.
    Kind regards,
    Jochen Klier
    National Instruments

  • PXI-7344

    Hello,
    I am using a PXI-7344 board with a UMI-7764 accessory. So I connect it to the motion control plug and it works properly (the 4 axes are all used).
    Next, I get a SCB-68 block connector because I need to get 2 more analog outputs.
    So, I use a cable SH68-C68-S to connect the SCB-68 block to the PXI-7344 digital I/O plug.
    But I don't know in fact how to steer the Digital Output ports from PXI-7344 to use the 2 analog outputs from SCB-68.
    Moreover, I would like to know if I can steer both devices from motion port and Digital port at the same time.
    Anyone an idea?
    Thanks

    Hello,
    I would like to measure some tests on charts with a PXI-4204 board and to save them in a binary file.
    I am using a VI that contains the "Analog 1DI32 Wfm NChan NSamp.vi" (from "Continuous Acq & Graph Voltage to File.vi") to acquire an analog signal.
    The problem is that I had to do a "mix" with other vis:
    - to be able to save the data into a binary file,
    - to be able to see the waveform moving from right to left and to use the X scrollbar.
    Now the problem I have concerns the reading.
    When I want to use the "Graph Acquired Binary Data.vi" to read the data, there is a message of error that says the input data is not incorrect....
    Anyone has an idea?
    Please see the 2 files enclosed: "Continuous Binary Acquired Data.vi" and "Read Graph Data Acquired Binary File.vi" (front panel +diagram)
    Thanks very much for quick reply
    Attachments:
    Continuous data acquired (diagram).jpg ‏162 KB

  • PCI-7344 motion controller encoder reset problem

    I'm using a PCI-7344 motion controller for my application. I'm currently working on simple display of position and velocity using an encoder. I am able to read the encoder pulser, but I am unable to reset it. I want the encoder pulses to be display continously till I press a reset button on the screen. After pressing the reset button on the front panel, the encoder count should become zero and start counting again till I press stop button. Kindly give me a simple example of encoder reset using push button. Thanks.
    Regards,
    Giridhar Rajan
    Cruiser Controls
    Mumbai, India

    This should do what you are looking for.  I am not at the computer that has a motion controller card in it, so I have not tested it. 
    It resets the encoders for Axis 1 and Axis 2.  Unless you have a different configuration for encoders and axes, it should work.
    Hope that this helps,
    Bob
    Bob Young - Test Engineer - Lapsed Certified LabVIEW Developer
    DISTek Integration, Inc. - NI Alliance Member
    mailto:[email protected]
    Attachments:
    Reset encoders.vi ‏66 KB

  • PWM PID loop on 7344 motion controller

    I need to control motion by PWM, with the help of 7344 motion controller.Please verify if it possible or not, and how can I do it.Please provide me(if it possible)any example of solution of this problem.

    Hello,
    The onboard PID circuitry can only be utilized with servos and closed loop steppers. You can write your own PID loop in software using the PWM functions and the Read Encoder Position.flx. This should work pretty well for low loop rates (since it's implemented in sw). I don't think there are any examples of it available but it should be pretty straight forward.
    Regards,
    Yusuf C.
    Applications Engineering
    National Instruments

  • Connect two 7344 motion controllers in one PC

    Can I use two NI - PCI 7344 motion controllers in one PC. My application requires 6 axis and hence I want to use two PCI 7344 motion controllers. I am already using one. Can I configure another one in the same PC to add more axes to my current system. If yes, then how to configure the other one, so that it acts as the 5th and 6th axes in my system. I'm using LabVIEW 7.1 in my application.
    Regards,
    Giridhar Rajan
    Automation Engineer
    Cruiser Controls
    Mumbai, India

    You can add as many PCI-7344 boards to your PC as you want. The only
    limitation is the number of PCI-slots available. In your software you
    need to control the axes from the second PCI-7344 seperately from the
    axes of the first one. That means that there are no built in methods to
    start the axes on two or more boards synchronously. If this is not a
    feature that you need then everything is quite straight forward. If you
    need tight synchronization you might write an onboard program that's
    running on the second board and that waits e. g. for a breakpoint
    signal to be generated by the first board immediately after the axes
    are started.
    Assumed that you need tight motion synchronization with the additional
    axes a better choice would be to use one PCI-7356 (6 axes) or a
    PCI-7358 (8 axes) instead of two PCI-7344. The 735x boards are pin- and
    software-compatible to the 734x boards. You could return your 7344 to
    NI and buy a 735x board with a fair discount ("hardware upgrade"). If
    this is an option for you please contact your local NI branch.
    Best regards,
    Jochen Klier
    National Instruments Germany

  • Problems with a PCI-7344 motion controller

    Hi all,
    I've run into a weird problem with a NI PCI-7344 motion controller, where I
    have a program that provides the motion control for my program that works
    reasonably well, but when I make minor changes gets motion errors. The changes
    are not specifically to the motion parts, just changing a couple of globals
    that provide communication with another part of the program (and another piece
    of hardware). The errors I get indicate that I am trying to enable limits
    switches at the wrong time in my program. I will add that I do not get any
    errors when I go through the program in debug mode
    Tnx,
    P.W.Monroe

    The portion of my program that gets the errors uses the "Flexmotion" Find Home
    and Find Index routines, which are built into the 7344 controller. In the
    exampes for using these VIs, they are followed by a loop that has a VI that
    monitors the status of these VIs for completion, both move complete and Home or
    Index Found. The globals that I mentioned are being used to terminate the loops
    if an error condition occurs elsewhere, and do not cause the program to do
    anything other than terminate looking at the status and then go into a wait
    loop. It crashes when it goes into the loop when I substitute a different
    global, which is baffling to me, as they aren't triggering something that would
    result in a race condition.

  • [Ask & Note] Feedback for BlackBerry 10

    Hi,
    I have some feedback for blackberry team but I don't know where to post it..
    So I'll just post it here and please let me know if this room is not the place for my post.. Thanks..
    I'm now using BB10 (Q10) and have some feedback for blackberry, so I hope this feedback can be used to improve BB10 features and also increase the easy to use and popularity of BB10
    Here is the list:
    1. Auto suggestion for email address (To/CC/BCC). Here I'd experienced that this features is useful and also very much annoying, why? because the list is come from the recent email address that we use to send and also from email address that have send email to us including the SPAM email (if coming to our BB Inbox), my suggestion is to make some feature/menu for user to have option for clean-up the list of email suggestion or user can delete some email in the list if they feel that email is not necessary to be there (like we will never send email to DoNotReply email from some company/domain).
    2. Recent Updates in BBM. I'm one of some users that not using 'Recent Updates' feature in BBM, and this feature is one of the reason why my BB battery drain so fast.. Correct there is option to turn off the 'Recent Updates' but the way is so much not user friendly, I must tap and hold each and every one in my contact list updates and hide their updates from my BBM, why not just put one option in setting for turn off ALL Updates (like u have in old BBM)? I think it will be more acceptable in this world compared to hide one by one (imagine if you have more than 1000 contacts in your BBM) and you must do it every time you do complete clean firmware upgrade.
    3. Mobile Network vs Wifi. I don't know this is just happen in my BB or other's too. My experience is, if I was connected to a wifi network and turned off my wifi, the connection supposed to be from Mobile Network, but what happened in my device is, the connection just shown E / 3G without BB logo (means no connection to Blackberry Service), it's happen until I manually turn off the Mobile Network and turn it on again or change my Network from 2G to 3G only vice versa. Please do check about this problem.
    so far that's the feedbacks from me, please let me know if there is some mistake in my post..
    Thanks a lot..
    I hope Blackberry will grown up again and more bigger than before..
    Best Regards and Thanks,
    Hartawan

    There is no link to suggest anything to BlackBerry but if any of the BlackBerry employees who browse these forums are going to look anywhere then it will be here...
    http://supportforums.blackberry.com/t5/BlackBerry-10-Smartphone/Feature-Requests-for-BlackBerry10-OS...
    If you've been helped click on , if you've been saved buy the app.
    Developer of stokLocker, Sympatico and Super Sentences.

  • Video Output for PXI Embedded Real-Time Controller ?

    I plan to purchase a PXI Embedded Real-Time Controller for security monitoring.
    Is it possible to use the video output of a PXI Controller running RTOS to display messages from a LabVIEW RT application ? According to the documentation, it seems that the use of the video output is only possible for PXI Controllers running Windows XP.
    Does it exist a specific application to do this with LabVIEW RT ?
    Thanx.
    H.L.
    Solved!
    Go to Solution.

    Hi Yann,
    The situation seems a bit confusing as in the link you've given above using the local monitor from a pxi labview rt system as the HMI without a pc attached, two messages (here and here) say that it is possible to display an image on the video output of a PXI RT controller using a function of the NI-IMAQ librarey (namely, the IMAQ RT Video Out VI) !
    I have the NI-IMAQ library but no PXI controller to test if this solution should work or not
    Does anyone have the answer ?
    Thanx,
    H.L.

  • How do I set-up my memory amount for max speed for FCP and Motion

    Hey all, I have a Dual 2.3 G5 with (8 Gigs) Ram. I'm working with a new client. I want to set up my memory or whatever else to provide MAXIMUM processing power for FCP and MOTION. Does anyone have any recommended settings. I've maxed-our RAM. My card is (ATI Radeon 9600-128MB of VRAM)
    Is there a setting in FCP that allows more processing power to be applied to that application ?
    Appreciate any replies

    Most of FCP's speed in performing functions is directly related to processor speed. The only way to increase performance would be to use a Mac with faster processors. Performance is also related to data throughput, so if you're working with high data rate video formats, a fast RAID and/or SATA (or eSATA) media storage solution would help.
    The exception to FCP performance is when rendering/processing FX Plugs; they are dependent on the power of the graphics card.
    Motion is mostly dependent on the power of the graphics card.
    There is no way that I'm aware of that you can speed up RAM. Having the extra RAM will just mean that you can successfully work with more applications open at the same time.
    -DH

Maybe you are looking for

  • Re: Display format in ALV report

    Hi Expert, I have develop an alv report with some like lifnr,name1,dmbtr,budat,zterm, But my requiremnt is output should come in the below format like vendor number   vendor name   payment tern     jan          feb                mar            april

  • How to get line in audio to internal speakers

    What procedure is needed to connect line in audio to the internal speakers?  For example, playing music from my iPad through the macbook Pro's speakers. Or from any other audio source,such as a USB connector?

  • IPhoto won't import photos from Digital SLR after Yosemite upgrade

    Has anyone else had problems with connecting usb devices and importing photos since the Yosemite update? iPhoto "sees" the camera, but it doesn't show on the desktop and never preview or loads the content of the camera in iPhoto. • MacBook Pro 13", m

  • Editing PDf's

    Hi, There is a pdf document i use as a template say, when I go to enter some text, into the name box for example, i get a drop down list of previous names, which is proving very annoying. Is there a way of preventing this from happening. Thanks

  • Will Microsoft Office For Mac V.X run on Lion?

    Need to determine if I have to buy the latest version of Microsoft Office For Mac.  Would prefer to use my V.X.  I am upgrading to Lion for iCloud.