Im peforming synchronized sampling from two boards; one is a PCI-6224 in my computer and the other a PXI-6229 located in our PXI-system. Being a rookie when it comes to labview, I was quite proud of myself for managing this all by myself. However, when i looked closer at a 10 sec test-record I noticed something peculiar. Although none of the 32 (differential) channels had inputs during the recording, the PXI-board returned 16 perfect 1 Hz sine waves with exactly the same amplitude as the measurement range. Attaching a signal to the terminal block didn't alter this. The PCI-board returns only weak 50 Hz (power network) noise, as expected. What is going on ??
I'm using a PXI-1031 chassis together with a PXI-8184 controller connected to the PC with a network cable. I trigger the pxi-board by sending the PC-clock through a PFI-port and a physichal wire.
regards, Einar

As I said, I haven't used LabVIEW 8 all that much. It's a bit tricky: you'll have to make a new Project, add the PXI as a new Target, and the add the VI you're trying to run underneath the new target. The last time I was playing with this I also had to open some ports in the Windows XP firewall.
If this is all a bit much for you, you might have a PXI controller that is configured to dual-boot both LabVIEW RT and Windows XP. If you reboot the controller in Windows XP mode, you can then access the PXI-6229 in the same way you currently access the PCI-6224... provided you don't mind copying the VIs over and hooking up a mouse, monitor, and keyboard to the PXI.
Switching from LV RT to Windows XP is most easily done by selecting the appropriate option in the PXI's BIOS. Check your PXI controller's manual for details.

Similar Messages

  • Generating a 6X Sine Wave using an absolute encoder

    How do I get an Analog Signal Generation Synchronized to six times the frequency of an absolute nine bit parallel encoders output with adjustable phase relation. I am using a PXI-6533, a PXI-6711 and have available a PXI-6070E.
    Now the long version of the problem. I have a 9-bit absolute encoder connected to a motor shaft (up to 420Hz). From the encoders 9-bit (0-511) parallel signal I need to create a synchronized 6X sine wave (up to 2500Hzwith an adjustable phase relation. Picture an electronic timing chain. For every cycle of the encoder, I need to produce six full sine wave cycles with a variable tuned phase shift. I have LabView 7 w/RT.
    I have tried to use the Update Clock (PFI5) to step through a buf
    fer but it seems to drop an update every so often. I also have tried to monitor the bit-0 line to determine freq. and scaled the update rate accordingly but the phase shifts all over the place. Does anyone have any ideas on the matter ? I would really appreciate it! Thanx….Phil

    Phil,
    Another different option to solve your application would be to use the new LabVIEW FPGA module and the 7831R reconfigurable I/O hardware. Using these tools you can configure the FPGA on the board to generate the sine wave based on the 9-bit parallel input. The most simple setup would be to have a lookup table in the FPGA that has a corresponding output value to each of the 512 possible input values. This 512 sample table would contain 6 cycles of a sine wave. Every time the parallel input changes the analog output value would be updated accordingly from the lookup table. This setup would also allow to easily add a phas
    e shift of the output by shifting the point in the lookup table you are reading by a known constant.
    Christian L
    NICS
    Christian Loew, CLA
    Principal Systems Engineer, National Instruments
    Please tip your answer providers with kudos.
    Any attached Code is provided As Is. It has not been tested or validated as a product, for use in a deployed application or system,
    or for use in hazardous environments. You assume all risks for use of the Code and use of the Code is subject
    to the Sample Code License Terms which can be found at: http://ni.com/samplecodelicense

  • Storing multiple Sine waves in one LVM

    Hi all,
    actually i got a very short question. I can't believe that i'm unable to fix the problem by my own... But that's the way it's suppose to be I think.
    The question:
    Just a simple Sine wave in a while loop, connected to a graph. When I run it i see a lot of different Sines in the graph window, let's say 100 sine waves. How can I write this data from ALL THE 100 SINE WAVES to 1 LVM?
    When i store the data right now I only see 1 Sine Wave and not all the waves i've seen in the graph window.
    I hope somebody can help me with this. Cause this is just a simple example of a very complicated programm i'm working on.
    Thanx for spending time on this stupid problem!

    Could you post your VIs?  I am not sure what you are talking about.
    Thanks,
    Bob Young
    Bob Young - Test Engineer - Lapsed Certified LabVIEW Developer
    DISTek Integration, Inc. - NI Alliance Member
    mailto:[email protected]

  • Generating a sine wave using PXI 5441

    How to generate a sine wave using PXI 5441 (Arbitrary Waveform Generator)?
    Can anybody post a vi that generates a sine wave with frequency at a user defined frequencies.
    Thanks in advance.

    Hi Venki,
     When using an arbitrary waveform or user-defined
    waveform, you provide the actual waveform you want to generate to
    the device, so it is easy to also just connect a graph to the waveform
    and see what it looks like. Unfortunately, when using standard function generation (which is the method used in this example), you are not providing a signal that is created in software. Because the standard function
    generation is happening on the FGEN and not in software, there is
    really no way to display the exact signal without measuring
    it on a scope or some device. You can, however, simulate the signal in
    software such that the parameters which are given to the FGEN functions
    are the same parameters that you use to generate a simulated wave. This
    would be somewhat complex since you can change parameters such as
    frequency and amplitude on-the-fly, and also you would need separate
    functions for each of the different signal types available for the
    FGEN. You would need to place the software waveform generator in your
    inner loop and have a waveform graph connected to it in the loop as
    well to display the signal. The main thing to keep in mind is that this
    will only be a simulated signal and so it may not exactly match what is
    being generated (this also depends on how you program the simulation to
    work). Hope this helps,
    Daniel S.
    National Instruments

  • PXI 5411 sin wave burst plus timing

    I have a PXI system with a 5411 ARB card and a 6259 MIO card.
    Here is the issue.
    I need to:
     generate a sin wave pulse at 16MHZ...
    This pulse initiates at T0 plus Tn from a trigger pulse generated by the AO of the 6259 (needs to be as short as 1mS and as long as 3000mS-repeatable and guaranteed)
    This duration of this pulse is Tx (can be as short as 1mS up to 1000mS again repeatable and guarenteed)
    My problem has been that the timing of Tn and Tx is at the mercy of the computer as ranges up to times as great as 1000mS.
    I know there is something simple that I am overlooking... but for the life of me, I have not been able to find it.
    Thanks,
    Paul
    Paul <--Always Learning!!!
    sense and simplicity.
    Browse my sample VIs?

    Hi Paul,
    I'd like to clarify a couple points in your post to make sure I understand your application correctly.  It sounds like you want to generate a trigger pulse on the 6259 - is this using the counter or analog output?  You also mentioned that you wanted your start trigger and reference trigger to be user-defined, but you are only getting 1000ms accuracy.  This may be occuring because you may be using software timing. 
    It sounds to me like you'd like to use a digital trigger from your 6259 to start your acquisition, then specifying the number of samples to generate to control your Tx time.  I may be missing something that you're trying to do in your application, so please reply with more details on your application if possible. 
    Thanks!
    Daniel L.
    Applications Engineer
    National Instruments

  • Simultaneous updation of sine wave using 2 PXI-6711 cards using RTSI Trigger

    Hi ,
    How to generate continuous sine wave using two pxi-6711 cards at same time without any phase shift using traditional nidaq functions.
    Regards,
    satya

    In LabVIEW, go to Help>>Find Examples. Navigate to Hardware Input and Output>>Traditional DAQ>>Multiple Device. There is an example there called "Two E Series Shared Scan Clock". That example is for analog input, but you should be able to modify it to be for analog output. Basically you will just have to share the scan clock from one board(master) and use it as an 'external' clock for the other board(slave).
    -Alan A.

  • I am trying to generate a 50kHz sine wave for ten cycles, then aquire this waveform and store it in a file, I am using a PCI-Mio 16E-4 board, has anyone done anything like this?

    I am trying to generate a 50kHz sine wave output on the DAQ card and then simultaneously recieve. I am using a PCI-Mio 16E-4 series board to do this. Has anyone done anything similar. I can generate a 50k wave no problem, and recieve one, but I can not get these vis to run together? Any suggestiosn

    Jrod, yes you should be able to do this, actually there are quite a few example programs come with LabVIEW that handles simultaneous analog input and output, I check the specs of the board you use, the board should be able to handle the sampling/generation rate, but make sure for 50KHz sine wave, you would sample at least twice as fast, preferably ten times the generation rate, then this puts the sampling of the board to thew limit.
    The program you can try is attached here, good luck with the application,
    XD Gao
    Applications Engineering,
    National Instruments
    Attachments:
    Simul_AIAO_Buffer(E-series).vi ‏104 KB

  • Recently i got my Macbookpro repair through apple service centre. As per them motherboard is required to be replace.I ask them after repair to return my old board-the service centre told me that apple will provided new board only on return of old one?

        Recently i got my Macbookpro repair through apple service centre. As per them motherboard is required to be replace.I ask them after repair to return my old board-the service centre told me that apple will provided new board only on return of old one? Is this the Apple policy, cn anyone pls help me out

    Hi p,
    Apple keeps all replaced parts. That's their policy.

  • PXI-5406: unable to generate 10 mV p-p sine wave with 1.5 V dc offset

    I'm using the PXI-5406 and FGEN Soft Front Panel to output a low amplitude sine wave. It works fine when I have the amplitude set to 10 mV and the DC offset set to 1V but when I increase the dc offset (I want it to be set to 1.5V) the FGEN front panel reports an error and stops outputting the wave. It starts working when I have the DC offset set to 1.5V and the amplitude increased to 50 mV. Any ideas on how to output a sine wave with 10mV amplitude (p-p) and 1.5V dc offset with the PXI-5406?
    Thanks,
    Rachel

    I also wanted to mention that while using the FGEN Soft Front Panel, some of the settings are coerced to set values (particularly Digital Gain). Is it possible for you to use the driver API through LabVIEW or another program? This would allow greater control and a potential workaround.(set Gain to a value less than one (0.25) and the Amplitude to a value to 40mV).
    Evan S.
    National Instruments

  • How can i measure only the valley voltages of a sine wave in labview 7 express evaluation version?

    I am trying to convince my company to send me to purchase labview and send me to classes for labview.
    I am trying to put together a small demo for them using the Labview 7 express evaluatin copy that I recieved.
    What I am wondering is how can filter out everything besides the valley peaks on a sine wave.
    What I am trying to accomplish, is create an example showing that Labview can sameple from a sine wave, read about 5 "valley" points of the wave, evaluate these based on a value that i can define, and signal pass or fail.
    attached is a picture of what i have so far
    Ryan
    LV 7.1
    Attachments:
    progress.bmp ‏3841 KB

    Hi,
    I built a VI using LabVIEW 7.0 that detects the valleys in a sine wave. It uses the 'Peak detector' VI that I believe is available in the evaluation version.
    (Functions palette >> All Functions >> Analyze >> Signal Processing >> Time Domain >> Peak Detector)
    I have attached a VI as well as the screen shots its front panel and block diagram.
    I hope this helps.
    Sincerely,
    Feroz
    Attachments:
    valley_detector_bd.jpg ‏36 KB
    valley_detector_fp.jpg ‏72 KB
    valley_detector.vi ‏29 KB

  • Synchronizing 2 sine waves

    I'm new to Labview and NI-FGEN so things are taking a bit longer than i think they should, the problem is im trying to synchronize sine waves from 2 PXI 5421 cards.
    what im fundamentally trying to do is have a slide bar and at one end of the bar one sig gen is at 5V and the other sig gen at 0.1V and when you slide it to the middle the amplitudes are both at 2.5v and at the other end of the scale the first siggen is at 0.1V and the second 5V.
    i can get this bit to work but when i connect the sig gens to a scope one signal is drifting.
    on another note if any one knows an easier way of creating the signals, I've "butchered" some examples. So there's probably loads of stuff that I don't need inthere.
    Attachments:
    test sine wave generator.vi ‏45 KB

    I do not have the FGEN VIs so my comments are general rather than specific.
    1. Unless the frequencies are EXACTLY the same, the signals will drift relative to one another. There are at least two frequencies involved for each generator, (a) the user-specified signal frequency and (b) the D/A clock frequency. Your front panel has two controls for frequency. If they are set the same, then the nominal signal frequencies will be the same. It is important to be very careful with fractional numbers because of the way they are represented in binary in the computer.  The clock frequencies will be nominally the same. Many devices use quartz crystal oscillators which have a frequency tolerance of +/-100 ppm (parts per million).  While 100 ppm is a small fraction, two signals gnereated by clocks which differ by that much will drift.
    Fixes: (a) Use only one frequency knob on the front panel and feed the value to both FGEN modules.
    (b) Use the same clock for both modules. I do not know whether these devices have a connection for an external clock. If they do not, you will never get perfect synchronization. Somewhere you will need to configure subVIs to tell the modules what clock to use.
    2. The Amplitude slider will be read only once at the beginning of your program. That value will be used throughout the run. Any changes made later will be ignored. This is due to LabVIEW's dataflow paradigm.
    Fix: The terminal needs to be indside a loop. Better fix: Learn about the Producer/Consumer (Events) Design Pattern. This will have a parallel loop with an event structure to handle the User Interface, in particular, the changes in the slider. It will send the new values to the FGEN loops via queues.
    3. Same as 2 for Frequency. 
    Lynn

  • How to measure time difference between zero crossing of a sine wave and rise time of a pulse on a same graph?

    I have a 50Hz sine wave and a pulse signal on a same graph. The phase difference between two is between 0-90 degrees.
    Now I need to calculate the time difference between (when the sine wave crosses zero volts) and (when the pulse rises). The frequency will stay approximately same for both signals.
    The application is for a three phase generator. In simple words, when the time difference between the zero-crossing of sine wave and the pulse rises increases, that means that the load on the generator has increases.
    I am a beginner user of LabView (version 9, 2009), maybe it is a very simple problem but I've been pulling my hair out for the last few days and coudln't figure anything out. Any help would be greatly appreciated. I am using DAQ USB-6008 to measure these voltages and pulse from the generator and a sensor
    I have attached a jpg (a graph that i just made with excel to explain). The time 't' is what I am trying to measure
    Cheers
    Awais 
    Message Edited by awais.h on 03-30-2010 11:20 PM
    Message Edited by awais.h on 03-30-2010 11:21 PM
    Solved!
    Go to Solution.

    Hi
    Thanks for the code but I'm afraid it won't work. Like you said the probability of choosing a value that is on both graphs may not happen. This is something that I would like the code to do automatically.
    But I did use the previous code posted and made some modifications to it and it seems to work perfectly. Now the next thing is to be able to get just that one value that tells you the first time difference.
    Here is what I get using that code.
    As you can see from the t Values. I only need the one that is highlighted. If there is a way to filter out the rest it would be great.
    I was thinking of a while loop and as soon as the value is higher than 3E-5 it would store the number and stop the loop, but I'm not too familiar with arrays in labview.
    Here is the the code modified.
    As you can see, it wasn't that big of a modification and it still is your code.
    I will keep trying.
    Thanks for the help
    Attachments:
    FinalShockSpeed.vi ‏55 KB

  • Outputting sine wave to motor -- While loop placement?

    I'm having trouble creating a basic Labview program to drive a motor with a sine wave. I'd like to output one sine wave at a time to my motor and look at the response to eventually get a Bode plot. I'm using the Sine waveform VI and I've got a PXI-7358 controller card.
    I'm having trouble figuring out what goes in the while loop. I'd like to output, for example, a 10Hz sine wave to my motor. But when I create a VI as shown below (& attached), my output waveform frequency is nowhere near my input frequency. When I type 10Hz I get no movement out. At 10.01 Hz input the motor oscillates at a few Hz, and at 10.1Hz input the motor oscillates wildly. I don't think it's a problem with my motor or my PID settings, because even when I eliminate the motor commands altogether (second figure below), I get the same unpredictable relation between the frequency I type in and the Waveform graph.
    It seems that the output waveform depends on some sort of difference between the while loop execution rate and the PXI-7358 command generation rate. It seems to me that the Sine Waveform.vi should be outside the while loop, but in that case the Load Position.flx only seems to grab one position. Is there a good way to have direct control over output frequency of the sine wave?
    Attachments:
    one_axis_move11.vi ‏20 KB
    one_axis_move12.vi ‏12 KB

    matthewg,
      Dennis is right, let’s keep this to one forum since it is the same topic of discussion, and I will address some of the issues with your code there.
    Chris_G
    Sr Test Engineer
    Medtronic, Inc.

  • Generating sine wave

    Hello,
    I have a small labview problem.
    I want to generate a sine wave pulse with labview and send this to a pc oscilloscope by using my soundcard.
    First VI I made creates a sine wave and sends it to the pc oscilloscope. Works perfectly.
    But now i want to create a sine wave with only one period (second VI). Problem is then that the VI blocks if i want a high frequency (I need 10kHz)
    So the VI works on low frequencies but gives this error with higher frequency:
    "the sound driver or card does not support the desired operation."
    Can somebody give a solution for my second VI (creating sine with one period on 10kHz).?
    Thanks in advance!
    Jelle
    Solved!
    Go to Solution.
    Attachments:
    GeneratingSineWaveManyPeriods.vi ‏59 KB
    GeneratingSineWave1period.vi ‏37 KB

    Hello Benje,
    The problem/difference you're seeing is indeed in the coding and also in the sound card.
    In the "working" example you're using the Simulate Signal VI with the following settings:
    - 44100 samples per second (sample rate)
    - Number of samples 10000 (samples per generation cycle)
    In the non working example you specify:
    - sample rate = 1000*"Frequency Control value". This sampling info should be fixed to (for example) 441000 Hz.
    - Number of samples 1000 (samples per generation cycle)
    As a sidenote:
    Is there any reason why you used different functions to generate similar signals in the 2 VI's?
    Kind Regards,
    Thierry C - Applications Engineering Specialist Northern European Region - National Instruments
    CLD, CTA
    If someone helped you, let them know. Mark as solved and/or give a kudo.

  • How can I get 'sync' signals for sine wave out? [FPGA USB 7856R]

    Hello,
    I'm Sean Ahn using NI 7856R, USB FPGA with LabVIEW 2014.
    I'm trying to make 'sync' signals for sine wave of adjustable frequency, like general function generator's sync out.
    I use sine waves whose frequency is feedback-congtrolled between 25~30 kHz and need sync signals of square wave.
    Like [1] below, when I try resettng two wavefom-generators, sine and sync signals are not perfectly synchronized.
    And everytime I adjust sine's frequency, the phase difference between sine and sync changes arbitrarily 
    In case of [2] using 'zero-crossing detector', I have severe jitter of sync signals not suitable as 'sync' signals for other parts.
    Please, give me some help.
    Thanks in advance
    Sean Ahn
    [1] Reseting two waveform generators
    [2] Zero-crossing detect

    Here’s my issue.  When I try to move any data between my PC and the iPad2 I receive an error message that states that “the [sync] session could not be started with the ipad, Please disconnect and reconnect the iPad and try again”
    I have tried again and again by various methods etc .  and all I get is the same message when I attempt to sync, backup or update the iPad. 
    I have tried the following AND FAILED:
    Deselect all Apps, Music etc in iTunes, then attempt to sync
    Disable and enable the iPad in device manager and restart itunes and try to sync again
    Disable and enable Apple Mobile Device Services and sync/backup/update
    Run iTunes as Administrator
    Reboot all devices (iPad buy pressing two buttons at once)
    Turn off all security and reboot and sync
    Connect directly to internet without a router
    Tried to sync with all wireless turned off and in airplane mode
    Running iTunes Diagnostics...... nothing all test passes
    Changing USB ports and cables
    Removing my SIM card and rebooting and trying to syc
    Trying to sync via wifi only
    Turning off all icloud services
    Uninstalling all Apple Software and reinstalling again
    I am syncing to a 64bit  PC running Win 7 Premium with all the latest updates, all firewalls off. WiFi and direct connect ethernet to Internet service with and without router.
    I am trying to Sync an iPAD 2 running iOS version 5.1.1 (9B206)
    I am trying to update from iOS 5.11 to 6.01 by using Itunes 11.0.0.16.3 in a USB or directly through the iPad via wifi.
    Please help as everything I have tried has failed.

Maybe you are looking for

  • My music won't play on my iPhone!!

    My Music just will not sync with my iPhone. A week or 2 ago I was make storage space and temporarily unselected the sync music library option. Then for some reason a lot of my document and data storage data was missing. from then on whenever I try to

  • Media Encoder CS6 won't launch - Mountain Lion

    After recently updating, Media Encoder will not launch.  I have tried searching for "watched folders" xml, which many say solved this issue for them on Windows, but cannot find any such file in ML. Any suggestions? I am running: Mountain Lion Mac Pro

  • Query about Load-Balancer 'proxy'

    Hi, When using load-balancer 'proxy', with multiple remote addresses defined, does the client randomly select the initial connection from the list of remote connections in the config file? I know the proxy will redirect a client to a less loaded prox

  • Connection between two or more outputs of PCI-DIO96

    Is it allowed to connect an output with low-Signal and an output with high-Signal of PCI-DIO96. In the manual I found only, that it is not allowed to connect +5V to GND. Thank you for your help Markus Pilhoefer

  • Error while upgrading 10g AS from 10.1.2.0.2 to 10.1.2.3

    Hi All, I am in process of doing 10g AS integration with 11.5.10.2. As per documentation i have seen that 10g AS has to upgrade My 10g AS version is 10.1.2.0.2 and my database version is 10.1.0.4.2 I have downloaded patch p5983622_10123_LINUX.zip and