I recently bought a RIO-7833R PCI card and am having weird problems with it.
I would like to use it to measure analogue voltages. However, even when nothing is connected to the card, the analogue input channels are railing to +/- 32768. I've written a very simple program to measure these voltages and it’s hard to see how there could be a bug in this code (see attached image). So, I'm thinking it has to be something with the driver of the card or some other bug in LabVIEW, or perhaps a setting on the card that needs to be changed.
Complicating things is that I actually had another card that was doing the same thing that I returned. When they sent me this brand new card it also gave the same values, so I do not think it is a hardware problem with the card. In addition, I've also changed the computer the card was installed in, and again, I get the same values.
Has anybody seen this before? Could there be a bug in the labview driver? I've tried changing the setup of the card in MAX so that the measurements are "differential” or “single ended,” but this does not appear to solve the problem.
Thanks,
Attachments:
FPGA problem.jpg ‏170 KB

Hi Adam,
I understand you are getting some strange readings from the analogue inputs on your NI PCI-7833R.
From what you say I agree that it is unlikely to be a problem with your specific piece of hardware. I have had a search to see if there are any known hardware faults for this device fitting your description and didn't get any results.
I don't believe the driver will be creating a problem for such a simple task as long as you have the correct one for your hardware and OS installed. You can check that it shows up in the software section in MAX.
I would suggest then that you try using the known working analogue input project for your device in the LabVIEW example finder which can be found on the start up screen.
In the example finder Select the Browse tab>>>Open the hardware folder>>>R Series>>> Basic IO>>>Analogue Input - R Series.lvproj. This has the FPGA VI to deploy to your hardware and a test Host VI. See if you can then get the expected 0 input when unconnected and test with a signal input on the graph.
Hope this helps
Jack. W
Applications Engineer
National Instruments

Similar Messages

  • Signal jitter in Analogue input using NI 9201

    Have a signal issue in the system we're developing. I am using several NI 9201 analogue input modules in a 4unit chassis connected to a 7841R fpga with a 8110 Pxi controller. The 9201s are being used to measure the output of several contactless rotary sensors plus some force transducers and other signals. All sensors have the same isolated DC supply and noise is relatively small. The first rotary sensor of the six however at infrequent random occaisions gets quite large jitter, about +/-0.2v at a random frequency and the voltage appears to increase gradually. Measuring the input signal directly with a multimeter does not seem to have the same jitter or rise in voltage and no other signals are effected. When the sensor is unplugged and replugged the signal is normal again or after the FPGA is reset. It probably only happens once every 8 or so hours of use.
    We had an identical issue with a previous system which used poentiometers instead of contactless rotary sensors. The only common hardware factor was that the problem only occurred on the first analogue input in a NI 9201 which is housed in the first slot on a 4u chassis. It is not the same 9201 module or chassis. Even the FPGA was a different model. All signals undergo the same filtering etc in the FPGA.
    Has anyone had similar issues or have any suggestions on how I should investigate/resolve the issue.
    Thanks in advance,
    Andrew

    Hi Andrew,
    Thanks for your forum post regarding the jitter you are experiencing with your analogue measurements using the NI 9201. Having researched your issue I have come across the help file article here which details a potential cause of jitter due to resource contention on board the FPGA.
    Basically the article describes that multiple requests on the same analogue input resource within the same FPGA clock cycle can introduce jitter, the jitter occurs most often when you use a shared subVI in two independent parts of the VI, or when you access a resource interface from parallel loops.
    Please let me know if this assists in the debugging of your issue.
    Many Thanks
    Jamie S.
    Jamie S.
    Applications Engineer
    National Instruments

  • Large analogue input noise with laptop and 6024

    Acquiring an analogue signal leads to large noise of +-100mV !
    I'm using a NI 6024 card with shielded cable and a selfbuild connector box on a Dell D800 Laptop. The analogue inputs are used as differential inputs and connected to external amplifiers. In the box additionally the frequency output (GPCTR0_OUT and DGND) is connected to a small electronic circuit with an octal buffer (74HCT250) and 5V voltage regulator (An external 12V power supply is additionally used).
    The noise is not reduced by connecting V_IN- directly or via 10k resistor to AIGIND.
    There is no noise, if a battery is connected to the analogue input.
    May I e.g. connect the DGND to the AIGND ?
    I would be happy to receive any suggestions for solving
    this problem.
    Thanks a lot
    Ralf

    Hi Ralf,
    When using laptops and external circuitry there's always a whole lot of noise generated by the conflicting coupling of the laptop's power source and your external source. You might find that using the external power supply but powering the laptop from the battery also helps. I'd suggest that first you make sure that your laptop and circuitry are powered from the same electric line, then you can connect your external digital ground to the AIGround via a 10K resistor, to protect the DAQ board. This will help, but I believe the best way will be to use a low pass filter between your amplifier and your Analog Input, that way you'll be able to get rid of most of the high frequency noise.
    In general, if you have an external digital circuit powered by an e
    xternal power source and you're using a laptop I recommend you use opto-couplers, powered by the 5V of the DAQ card to get rid of any possibility of noise.

  • Inspire 5600 5.1 speakers, not enough analogue inputs on receiver? help!

    Hi, I need some help connecting my computer up to my Inspire 5600 speakers. I want to set up an analogue 5.1 connection, as apparently surrouund sound in many games doesn't work with optical. My problem is with the inspire 5600 receiver. It only has analogue inputs for front and rear speakers, and none for centre/subwoofer. I tried using the Creative technical support site, but for some reason my speaker set isn't listed (it has the Inspire 5200, and the I-trigue 5600, but no Inspire 5600). I checked the guides for other speaker sets to see if they were relevant, but the flash animations all show 3 analogue inputs, not the 2 that my receiver has. My speakers are a few years old, and unfortunately I don't still have the documentation that came with them.
    My question is where do I plug the centre/subwoofer? Does the Inspire 5600 not actually do 5.1 for analogue? (Aside from the analogue front/rear inputs, it has an optical in, digital din and a coaxial). Does the centre/sub maybe plug into one of these instead? I'm very confused, any help would be appreciated.

    Sorry I mistook it for another model, my bad. The Inspire 5.1 Digital 5600 has only two analog inputs for front and rear channels respectively. You are right, there is no orange jack on the subwoofer for this model. Last I checked, I was told that this speaker model was released before sound cards with 5.1 connections came about. So if you are doing up an analog connection, you will lose the center/subwoofer channel. However if you own a Creative sound card, you might be able to enable the CMSS feature to create a virtual surround.

  • Recording from analogue input

    using cool-edit or audacity. recording through mic input/output on left panel results in feedback howls from internal mic still active.have tried all combinations in sound manager settings.
    can i record through usb input with phono to usb cable or do i need a analogue to digital converter?

    Hello @8323,
    Welcome to the HP Forums, I hope you enjoy your experience! To help you get the most out of the HP Forums I would like to direct your attention to the HP Forums Guide First Time Here? Learn How to Post and More.
    I have read your post on how you are looking to record from an analogue input using recording software. I would be happy to assist you in this matter!
    Are you recording audio with an analogue device while your internal microphone is still enabled in the sound manager? Is a howl sound made when you are recording with only the built-in microphone? In the meantime, I recommend following the steps in this document on Resolving microphone problems (Windows 8), which should help enable your system to record a better quality of audio.
    Please re-post with the results of your troubleshooting, as well as the requested information above. I look forward to your reply!
    Regards
    MechPilot
    I work on behalf of HP
    Please click “Accept as Solution ” if you feel my post solved your issue, it will help others find the solution.
    Click the “Kudos, Thumbs Up" on the right to say “Thanks” for helping!

  • Synchronize analogue inputs with digital inputs

    Dear fellows
    I am using NI-DAQ, PCI-MIO-16E-4 (new name is 6040-E, I guess) with LabView 7.0. I am working on neuro-physiological signal acquisition. I have 16 analogue input channels and one digital input channel. Digital input is acquiring a TTL level signal that is a reference to the onset of a particular event. I want to synchronise both analogue and digital signals.
    Any suggestion?
    Nabeel

    Nabeel,
    have a look on following KB:
    http://digital.ni.com/public.nsf/websearch/851A80A7AB2885E086256F4D000610E6?OpenDocument
    (it's similar with DI)
    regards,
    ThSa
    http://www.newgistics.com

  • I'm not able to set Analog Input voltages, etc

    ok,
    so i have an NI DAQ I/O card (NI PCI 6723, or something to that effect) and i'm trying to configure the channels for Analog Input signals.
    when i try to choose any kind of setting for Anolog signal by:
    >>drop DAQ Assistant vi onto the block diagram
    >>>>choose Analog Input
    >>>>>>>Voltage
    i get this message "No supported devices found"
    i thought since I have the I/O card, i should be able to configure channels for both input/output.
    i also have ADAC card (from IOTECH) installed but when i look at MAX, i only see the DAQ card.
    i'm really confused at this point. can anyone break things down for me ... what i'm doing wrong?
    -r

    stuartG,
    you were right. so what i'm using for the input is a ADAC 5501MF from IOTECH
    following their manual (http://www.iotech.com/productmanuals/adac_lvi.pdf) it says that the ADAC-LVi libraries can be located at "Functions->User Libraries" palette of labview.
    does this mean that i have to install the ADAC-LVi driver in LabView->User Libraries directory?
    if that's not the case then i don't see why the ADAC-LVi is not showing up in the LabView User Libraries since i've installed the driver correctly (in C\Program File\ADAC )
    do you have any idea why?
    thanks
    -r

  • How to address digital input & output, analogue input & output when communicate with mitsubishi fx1n plc

    dear all,
    i am new to labview, how to code or address digital input & output, analogue input & output when communicate with mitsubishi fx1n plc or crouzet plc.
    pls help.
    HM

    http://search.mywebsearch.com/mywebsearch/redirect.jhtml?searchfor=mitsubishi+fx1n&cb=ZO&p2=%5EZO%5E...
    Try the above link. I know it is a screen-full, but is a search result for Matrikon. They have a whole slew of OPC servers/clients you can try out. Personally, I would go with their OPC server for the Mitsubishi, and use Labview Datasockets. That would be the simplest with the lowest learning curve.
    Otherwise, if you wish to use Modbus, you would have to get the commands/syntax from Mitsubishi; might be in their user manual, but I wouldn't lay bets on it. With the Matrikon stuff, the syntax is all built in.
    Good luck which ever way you decide to go.
    Dave

  • Varying input voltage refrigerator testing application

    I am working on an application for refrigerator testing. One of the specs that we need to comply requires that we vary the input voltage, (ex. 127 VAC for a 120 VAC unit) through some sort of voltage regulator. The biggest unit we test has a 2 hp compressor with a max current draw of 18 amps. We would like to know what equipment we would need in order to accomplish this through a LabVIEW application that we are working on. We could do this through GPIB, serial, etc. Any help on this matter will be appreciated.

    You might be right.  I was think of one small 1.5 HP controller (Commander SE) we are using that is being powered by 1-phase 240 VAC supply.  So I was thinking the motor was 1-phase, but actually it does the voltage chopping to create a 3-phase supply to the motor.
    It may be possible to still use one where you just use 1 phase relative to ground, but I'm not sure, and I'm not an electrical engineer to be able to tell how to do such a thing.  It may be worthwhile to ask them if such a thing can be done.  Or look around for some other motor speed controller manufactures that make a 1-phase motor controller.
    I'm sorry, but I don't know where else to look or be of any help.

  • Input voltage on hub

    I have a Linksys EFAH16W work group hub I wish to use, however the power transformer is missing. I find no markings upon the case or circuit board stating the input voltage value or, if its AC or DC, from the transformer.
    If you own one and can spare a moment, look at the transformer and find the output listings and share them.
    Thank you very much..

    Can anyone tell me where I can purchase one of these. I have had problems locating one.

  • Error when trying to edge-trigger to a input voltage smaller 300mVpp

    The programm shown in a cut below (just a cut of it) reads multiple channels from the PXI5105. The program allows to choose between different triggers. When I choose edge-triggering and the value of the input voltage I am triggering to is smaller than 300 mVpp the program stops and sets an error. With the 'Highlight function' I found out that the error is set at the '1D WDT'-vi.
    The error message:
    ERROR
    "Possible reason(s):
    Driver Status: (Hex
    0xBFFA4009)
    A previous acquisition is still in progress. If you are
    attempting to change an attribute, note you can only change fetch attributes
    while an acquisition is still in progress.
    Status Code: -1074118647
    I guess, if the voltage gain isn`t high enough the '1D WDT'-vi waits the 'timeout' = 5s for the trigger, isn`able to get one, and stops. But what can I do to trigger to smaller voltages as 300 mVpp? There is no option to change it. The trigger level of the 'Analog Edge Ref Trigger' is set to 0.
    Attachments:
    error1.jpg ‏120 KB

    Well then that's not it.
    Not triggering at low voltages- can be caused by either scale too high
    (so that 1 bit is bigger than the trigger level, not the case here) or
    scale too low (because of saturation the acquired signal never reaches
    the trigger level).

  • How to plot two analogue inputs from a DAQ device against eachother on an XY graph

    I am trying to separate data I am acquiring from two analogue inputs on my NI USB-6009 in order to plot them againt one another on an XY graph. I have tried setting up two DAQ assistants on the block diagram, one acquiring data from ai0 and the other from ai1. I tried to wire these to the X input and Y input respectively and running in a while loop but Error -50103 shows up.
    I've read through a few discussion threads and have found that you should acquire the data from both analogue inputs thorugh one DAQ assistant, useing index arrays to split the data and wire to the X input and Y input of the XY graph, but I cannot figure out how to do this.
    Any help would be greatly appreciated!

    It's pretty trivial to use a single DAQ Assistant. When you create one, you are given the option for multiple channels and shown how to select them. Once created with a single channel, you simply open the assistant and click on the Add button. After the acquisition, you use the spilt signal function if you stick with the evil dynamic data.

  • Noise at analogue input of PCI-7350

    Hi,
    I have a set up as shown below. 
    Futek Load Cells >> SCC-SG04>> SC-2345 >>SH68-68EP Cable >> SCB-68 >> Analogue input of PCI 7350 8 axis motion controller.
    This set up has resulted in a noisy analogue signal from 7350 with high amplitude spike. 
    I had try connecting the signal out of SCB-68 directly to an oscilloscope, and the signal was fairly clean without the spikes. however, when i connect the signal to AI of PCI-7350, both oscilloscope and load ADC.vi readings of the signal became noisy with high frequency noise on oscilloscope. All the AI ports and AIGND ports were probed without other connections showed the same noise problem.
    Moreover, after the switching on of 24Vdc power supply for the motors, the amplitude of noise became even higher.
    Is there any method to eliminate/minimise the noise pick up at the AI ports?? Any suggestions are welcome.
    Thank you.
    Yanxinix 

     An essential issue when connecting analogue lines to a system is connecting not only the signals lines properly but also the ground lines. Proper grounding of the signal connections is essential. You should be aware that BOTH lines (signal and GND) carry the signal, and depending on the connection of the GND line excessive current flowing through this line on a section between the system and the sensor may cause excessive noise. Much has been written about proper grounding of analogue signals. The most essential rule is NEVER to let any significant current (besides the current generated by the signal source) flow somewhere between the system and the sensor, i.e. NEVER use the signal GND lines as GND line for the power supply of the sensor. There are tutorials available on this issue on the NI site, try this one: http://zone.ni.com/devzone/cda/tut/p/id/3344.

  • Why input voltage couldn't display in chart ?

    In my vi,why input voltage couldn't display in chart ?
    please see the attachment
    Attachments:
    singal requency.vi ‏33 KB

    Well, your ai runs at 10000Hz and your ao runs at 20Hz, so start there .
    If they should be really synchronous, run them off the same clock.
    LabVIEW Champion . Do more with less code and in less time .

  • Problem with analogue output voltage

    i am working on a level control system which has an analogue input (0-10V level measurement probe) and an analogue proportional valve analogue output. I have configured the AO to be 0-10 volts in the measurement and explorer but when running the labview program it reverts to -10 to 10volts.Therefore if i set a tank to be filled to a setpoint of 3 volts it will do this but if i set the set point back to 2volts when the probe is reading 3V the Analogue output reverts to a negative output and opens the valve to fill the tank even if the setpont is reached.
    Can anybody help?

    Hi, thanks for your reply. The card is a PCI-6024E type the version of labview is Labview 6.Enclosed please find my labview code and front panel as an attachment.
    Thanks
    Micheal Hennessy
    Attachments:
    PID_auto_Control_Loop.vi ‏70 KB