I am using LV5.1 and a DAQ-card Computerbo​ards "CIO-DO96H​" (DO-card) since years.

Is it possible to readback the outputstatus in LV 5.1?

Chris:
The answer to this question depends on the hardware you are using and also the driver software that LV is interfacing(supplied by Computerboards). The best thing is to check with them to see if the hardware and device driver supports readback functionality on the output ports.
June

Similar Messages

  • Can I use ARM LM3S8962 like a DAQ card?

    I have the ARM embedded software and have used it to write some programs that run on the LM3S8962.  What I'm wondering is if I can have a program that runs on the PC that "talks" to the ARM board occasionally to query the inputs or control outputs.  If it is possible, how would I do that?

    You absolutely can use an ARM LM3S8962 like a DAQ card, the only restraints you have are how fast the microprocessor can run and the limited amounts of AI you have available on the Luminary board. you can simply add an AI elemental I/O node to your project and wire your transducer to the luminary board. You then can take the data from the microcontroller and send it to the Host PC through a network published shared variable. There are a couple of tricks to this that may not be completely evident. I have included two links that I think will help you get to where you want to be. The first is setp by step how to connect an accelerometer to teh ARM board, but you can modify it for your own transducer.
    http://decibel.ni.com/content/docs/DOC-9614
    The second is hwo to use network published shared variables on an ARM board. I hope this helps!
    http://digital.ni.com/public.nsf/allkb/6E50B28DDC61C0E68625767400805AA0
    Scott W
    National Instruments
    Applications Engineer

  • Modbus protocol and NI DAQ Card

    Hi,
    I'm using an AC drive that supports the standard protocol Modbus RTU over RS485. I want to know if your data acquisition cards (eg. NI DAQCard-6036E for PCMCIA) are suitable to interface between this AC drive and a computer. Also, I will be using Labview to control and monitor the system.

    Hi what,
    If your AC drive uses the Modbus protocol over RS-485 then you will probably want to use the serial port or an RS-485 port to communicate with your AC drive from LabVIEW. I believe Lookout also comes with an OPC server which can communicate Modbus. Anyway, I suppose you could use the digital lines of the DAQ card to control your drive but it sounds like an RS-485 port (serial) would be more effective and easier than programming the DAQ board.
    Anyway, hope that helps. Have a good day.
    R

  • Read/acquire signal from remaining empty channels of NI9205 by modifying the vi of ATI Force/Torque Transducer which uses the first 6 channelsof DAQ card.

    Hello,
    I am using force/torque transducer by ATI Automation. It provides me with its own vi to measure and write the data of the 6 signals (3 axes Forces, 3 axes torques)
    Following is my Ni system: -
    cDAQ 9172 chassis: Slot 5 - NI 9401, Slot 6 - NI 9205, Slot 8 - NI 9237. (Excitation voltage to the sensor/transducer (0-5V) using 9237).
    I connect the 6 signals from the transducer cable to channels ai0,1,2,3,4,5 in NI 9205(differential). The vi uses a calibration 6*6 matrix to finally display the calibrated voltage data.
    Now, I wish to use the remaining availabe channels that empty in 9205, namely ai17, ai18, ai19 for other signal measurements.
    ai17 channel is connected to Hall sensor  that gives out a square waveform corresponding to the rotor rpm. ai18, ai19 is connected to rotor-motor power supply's voltage and current signal respectively.
    But I am not able to access these 17, 18,19 channels from ATI company's vi. The vi loads the complete DAQ card (9205), uses only first 6 channels, hence the array data wire consists of only 6 channels. I can split these into 6 individual signals only.
    If I use a separate DAQ task to read channels ai17,18,19, then I get the error that this channels are reserved for some other task. Please tell me how can I access all the remaining channels apart from the first 6.
    I am attaching the vi here.
    Steps to run the "MEASUREMENT main.vi" : -
    Load calibration file:-  FT8840.cal;
    Load DAQ card: NI 9205
    Load ctr0 of NI 9401 (I am using this counter for rpm measurement purpose, this is my addition to the actual vi) 
    Attachments:
    DAQ - Copy.zip ‏574 KB

    There's nothing I can do to help.  I don't have your hardware to be able to modify your code and to set it up to make sure it runs properly.  Most people on the forum probably don't either.  This is where you'll have to put your programming and LabVIEW skills to work to solve your problem and make it run the way you want to.  If you run into a specific problem and get stuck, then please post back.

  • HT204266 I used several apple and iTunes gift cards on my account with relevant codes.  How can I verify all cards have been used when total purchases don't line up?

    I have used several apple app and itunes gift cards to purchase apps and music with no problem.  Codes were entered as required.  Is it possible to get a printout of the card amount and codes to check against purchases.  My total is not jiving.

    No, you can't see a list of card redemptions, nor how items were billed - all you can do is view your account's purchase history via your computer's iTunes : log into your account on your computer's iTunes via the Store > View Account menu option, you should then see a Purchase History section with a 'see all' link to the right of it. Click on that and you should then see a list of your purchases.

  • Why do error comes when trying to use NI ISA GPIb and NI DAQ card (both PnP)together while they work well when one at a time?

    we configured gpib pnp and at-mio-64e3 cards (isa) in the same pc.gpib function was invoked to take the device to remote and listen mode.now when daq function getDeviceInfo is called the remote mode of the device is removed and after this it always shows no gpib board until pc is rebooted

    Hi,
    I have some questions:
    What OS are you using?
    What driver versions?. You can find the driver versions in the Windows Add/Remove programs utility. DAQ driver is NI-DAQ anmd GPIB driver is NI-488.2.
    Is there any error code?
    I haven't found any known issues between these two drivers. How are the board's resources configured? Base address, IRQ, DMA.
    Let see if with this information we can clear out what is going on.
    DiegoF.

  • Time jumps when I use NI 9217 and the daq assistant

    Hi,
    in my program I want to measure temperature and pressure,
    the pressure part have no problem and I get the right signal in the right time but if I start the temp. Measurement I get time jumps in my graph and I have no idea why this happen.
    I have tried every setting from the daq ass. for the measurement of the temp. (sample, rate, timeout) but it makes not really a difference.
    Pleas help me before I freak out.
    Best and regards
    sascha
    Attachments:
    3druecke_temp.vi ‏165 KB

    ok the problem is your " time has elapsed " vi . Try the attached vi . remove the case structure and you can see the difference . it will give you an idea ............
    Attachments:
    difference.vi ‏68 KB

  • Can I use my AT-MIO-16X DAQ card on a PC running Win2000

    I am currently using an AT-MIO-16X card, on a WIN98 PC with Labview 5.1 and the SQL Toolkit. the network admin folks are wanting/demanding I upgrade to Windows 2000. Will this old card work on Win2000?

    Hello Hjs,
    AT-MIO-16X DAQ board will not work with Win2000. You find the following link useful in your search.
    http://www.ni.com/support/daq/versions.htm
    Shruti T
    National Instruments

  • How to communicate with a 9401 module using 9151 expansion and 7813r PXI card

    Hi,
    I have just started working with my 7813 R-series card.  I was able to create a simple VI in which I could send and receive digital signals.  Now I would like to do something similar using my 9151 expansion chassis and 9401 C-series module.  If possible I would like to do all of this using windows (no RT for now).  Right now I am right-clicking on "my computer" in the project tree and adding new devices to find the 7813.  From there I have a rough idea of how to build a simple VI using the NI example.  However, I do not know where to look for the expansion chassis or the C-series module.  If any one could give me some advice or point to an example, that would be great.  I have seen the examples in Example Finder for the 9401, but they seem to be using a compactDAQ system as opposed to the PXI setup. 
    Thanks for the help!
    Solved!
    Go to Solution.

    Hello ACE_Lab,
    The steps you should take in order to configure your setup in the LabVIEW project window is as follows:
    Step 1...
    Step 2...This step should auto-populate with the existing PXI-7813R if it is connected. If you want to setup the target without the card connected, the you can just add a new target as follows.
    Step 3...This is the point at which you will be adding the NI-9151 expansion chassis:
    Step 4...
    Step 5...
    Step 6...If you have the module connected to the chassis, you should be able to see the module in the list:
    Step 7...If the module is not connected, you can just select "New Target or Device" from the menu in step 6 and add a new target as follows:
    An example from the Example Finder called "Digital Line Input.vi" should be a good place to start. Below is a screenshot of roughly where you need to start:
    Regards.
    Jeff L.
    Applications Engineer | National Instruments

  • Using Ethernet network and Wireless to connect to ARD

    Hi, I am using ARD to connect 2 mac minis, 1 macbook pro and a G4 powerbook. I have been having some serious issues with trying to network them all together. When I connect 3 of the machines using ethernet and an airport extreme everything is fine, but the 4th machine is uncontrollable. It shows up on the scanner but shows up as offline. Has anyone had any luck using a combination of ethernet and wireless? It is all through the same network, but the hope is that the 2 laptops can connect wirelessly to the mac minis.

    MATT BANEY wrote:
    No my DSL modem is not a dumb one, its a DSL modem router, "qwest Actiontec dsl modem wireless gateway", GT701.
    To avoid confusion in future support threads, you should refer to that as a "router" or "modem/router", rather than simply as a modem.
    This configuration works fine, both computers are on the network, get assigned different IP addresses by the DSL Modem, and most of the time can see each other and the internet just fine, as long as the iMac keeps the Network settings that I set, for the IP address of the DSL-Modem router ("xxx.xx.0.1"), but occassionally after a restart it will revert back to using "xxx.xx.1.1" for the router IP address which is not a valid IP address on my LAN.
    Why would the iMac try and use the old settings when I specifically set and LOCKED the correct settings?
    If the "modem router" is configured to be a DHCP server and if your computers are configured to get their network settings via DHCP (Network panel of System Preferences, "Configure IPv4" set to "Using DHCP"), then in theory you shouldn't have that problem.

  • Can my phone be used if it reset using recovery mode and the sim card is taken out and the phone is network unlocked? (iphone 5s)

    If someone steals my Iphone 5s, removes the sim card, leaves the phone in airplane mode, puts in it recovery mode and resets it completely and deletes everything, then gets it network unlocked at a shop, and uses a pay as you go sim in the phone on a different network, can it be blocked or used by them still on a different network? I live in the u.k, and want to know what the thief can do with it? Can it be blocked if the thief does all this? Thanks.

    Okay thanks, what happens if the apple id on the phone is an old itunes account in which you no longer have access to the email adress associated with it and you've forgetten the password? And lets say for instance you never used a credit card on the account and you filled out the details like the date of birth wrong when making the id? There would be no possible way to sign in to activate it and no possible way to recover the apple id. So what would happen?

  • HT1918 Someone is using my account and charging my card that I have not agreed to

    I have 6 charges to my credit card in the past week, that I Did not agree too and I would like, the monies returned to my bank account.
    1) 138050473014
    2)  120050283367
    3)  127050285928
    4). 136050264270
    5). 183050125586
    6). 115050619256
    7). 127050230748
                                            Thank you.   Burt Gunn
    <Email Edited by Host>

    These are user-to-user forums, you are not talking to Apple here - I've asked the hosts to remove your email address from your post.
    You can contact iTunes support via this page : http://www.apple.com/support/itunes/contact/ - click on Contact iTunes Store Support on the right-hand side of the page, then Purchases, Billing & Redemption

  • Problems using well-configured ADLINK PCI9111 DAQ card

    "Hy
    I don't know why but my ADLINK PCI9111 don't work with LabView 6.0. It is recognized in Windows2000, and I have installed the PCIS-LVIEW driver from Adlink. When I run a VI taken from the exemples of this driver to generate a waveform, I have an error message. Can you help me ?"

    Hello Pierro,
    Unfortunately I do not have a lot of experience working with ADLINK boards. Is the error you are getting generated by LabVIEW or ADLINK driver software. If it is generated by ADLINK's driver software, please contact ADLINK support Engineers. However, if the problem keeps occuring after you have taken out the ADLINK board, please post the question with the LabVIEW error code and we will be more than happy to help you.
    Thanks for calling National Instruments
    Serges Lemo
    Applications Engineer
    National Instruments

  • My version of Firefox is sooo slow in response. I'm using OSX snow leopard. Should it take 3 or 4 seconds to switch tabs? it didnt use to. And loading hotmail pages is worse. This compares to previous years when it was the fastest browser out there

    Mac mini 2.1 Intel core 2 duo 1.83ghz
    Mac OSX 10.6.8
    FIreFox 9.0.1

    Create a new profile as a test to check if your current profile is causing the problems.
    See "Basic Troubleshooting: Make a new profile":
    *https://support.mozilla.org/kb/Basic+Troubleshooting#w_8-make-a-new-profile
    If that new profile works then you can transfer some files from the old profile to that new profile, but be careful not to copy corrupted files.
    See:
    *http://kb.mozillazine.org/Transferring_data_to_a_new_profile_-_Firefox

  • How to generate a pulse waveform 90 deg phas shifted with respect to a pulse generated using a counter in PXI 6070E daq card ?

    Hi
    I ma using 6070E daq. I am generating a infinite pulse train using one of the counters. I want to generate another pulse train which should 90 degress phase shifted with respect to the previsously generated pulse. How to implement this using the counters in the DAQ card.

    Hi Gopal,
    1) yes the frequency generator is programmable. You have 2 timbebases (10MHz and 100kHz) and you can divide down by any integer between 1-16. Those are the only frequencies you can use.
    2) The way counters work is that you can apply a gate signal to the counter which will cutoff the output. Therefore, you can have a second counter generating a pulse train of an equivalent frequency to the frequency generator (fout) but you don't want it to start immediately. You want it to start only after a short delay (equivalent to a 90d phase). Therefore, you need to "gate" or prevent the output on this counter for that small delay period of time. The way you can do this is by using a second counter to create a delayed pulse. You would
    connect the output of that counter to the gate of your pulse train function. That way, during the delay period of the pulse, you won't be outputing your pulse train yet. When the pulse (high-time) of your pulse reaches the gate, then your pulse train will start generating its pulse train (which is consequently delayed by the amount of delay used in the single delayed pulse counter). Since you don't want your pulse to go back low, you will have to turn off or clear your pulse counter while it is high. That way the signal on the gate will remain high forever.
    Your better solution is to use a PCI-660x card to perform this action. You will have more counters which will give you a greater range of frequencies to choose from and you will have more accurate timing of the signals.
    Anyway, hope that clears things up. Have a good day.
    Ron